AMX 4 Error Codes

Tom Henderson

AMX 4 Error Codes

January 10, 2005 11:38

I have an error 439 on my AMX 4, I cannot find a listing of error codes in our service menu. Does anyone know what this error is and how I could get a copy of the error codes. Thanks.

Log inor Register

to rate and post a comment

(1)

Tom Averbeck

Re: AMX 4 Error Codes

January 10, 2005 12:51

AMX 4 Error-439
Generator Fault
Probable Cause-The X-Ray On status signal was not asserted within 2ms after the Exposure start command signal was asserted.

Recommended Action-Check "EXP START CMND" signal of the Generator Control 1 port on sheet6,locationB7, and "X-RAY ON" signals of the Critical Status port on sheet3,locationD8 of the CPU schematic.

Log inor Register

to rate and post a comment

(110)
(10)
(3)

Deon Armes

Re: AMX 4 Error Codes

January 17, 2005 12:54

There are 23 things to check in the error code listing for this error. email me at axsmedical@yahoo.com we can discuss them.

Log inor Register

to rate and post a comment

(1)

Gordon Stokes

Re: AMX 4 Error Codes

March 10, 2005 11:29

Did you ever get a listing of all of the error codes for the GE AMX 4? I would be interested in such a list myself.
gordstok@sarmc.org

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats

Événements de Vente aux Enchères à Venir

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)