need help for Ziem 8000. System comes up with
Error E61. Any one can provide us with service
Manual or technical support?

gunter braun
medisys-braun@t-online.de

Log inor Register

to rate and post a comment

(1)

Simon Pritchard

Re: Ziem 8000

November 09, 2004 02:46

E61 = CCD camera iris problem.
Could be a faulty potentiometer on the iris assembly,
or a broken cable back to the uP board.
Hope this helps.

Log inor Register

to rate and post a comment

(2)

Krikor Najarian

Re: Ziem 8000

November 10, 2004 04:40

IRIS problem.

Check potentiometer/ iris motor / the mechanics of the iris.