Over 500 Total Lots Up For Auction at One Location - NJ 06/15

XR6000 error 020-0253

Tags: Radiology - R/F, GE

(4)

ENRIQUE RIOS

XR6000 error 020-0253

June 05, 2020 09:41

Hi Greetings.

I am working in a XR6000 the equipment is not generating x ray, the system turn on, but no emision of x ray, the anode is not rotating, and only appears the error 020-0253, and sometimes the error 020-0212, some suggestion?? my experience in this equipment is poor.
Thanks in advanced.


Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: XR6000 error 020-0253

June 08, 2020 11:10

Code: 20
Heater Warning

0253H
No CAN message received within 4 secs

code 50
50-0212H
MAINS_DROP detected.
The firmware of the Heater board has detected the mains_drop signal activation and has transmitted error to kV control
1.Interference (spikes)
2.Mains drop
3.Cable or connector contact in DC bus between AC/DC and auxiliary

Log inor Register

to rate and post a comment

(1)

KOHOUNKO ULBRITCH

re: XR6000 error 020-0253

June 10, 2020 11:16

check the anode board