I have a problem in ValleyLab force FX electro-surgical unit. Cutting display shows Error code 212. I will be grateful if some one can inform me that how this problem can be solved.
Dear Sir,
I replaced the battery but the problem 212 is still appearing. I dont know who can I set the date and time?
I need service/instruction manual.
My email is dmesf04@yahoo.com
212 Generator not calibrated Refer to Calibrating the Generator on page 5-26.
Calibrating the Generator
There are eight calibration steps. During calibration you verify information specific to the
generator, adjust the date, and adjust the clock. You also adjust the REM circuit and
several values, or factors, that ensure the proper operation of the generator.
The calibration steps and the values you can adjust are summarized below. Certain values
are not adjustable but must be verified.
Notice
After completing any calibration step, proceed to the next step to save the values from the
completed calibration step.
Calibration Step and Description Adjustable?
1. Generator data
Generator model number
Master microcontroller software version
Feedback microcontroller software version
No (verify value)
No (verify value)
No (verify value)
2. Calendar
Month
Day of the month
Year
Yes
Yes
Yes
3. Clock
Hour
Minute
Yes
Yes
4. REM oscillator frequency and impedance Yes
5. Current sense gain factor Yes
6. Voltage sense gain factor Yes
7. Reactance gain factor Yes
8. ECON factor Yes
Preparing for Calibration
Equipment required
• Bipolar footswitch and monopolar footswitch
• Two small test cables (less than 24" [61 cm] long) with banana plugs
• REM plug (modified)
• Resistor substitution box
• Oscilloscope
• True RMS voltmeter (such as the Fluke 8920A or equivalent)
• Current transformer (such as the Pearson Model 411 or equivalent)
• Noninductive power resistors (such as a Dale NH-250) with these values (in): 10,
20, 30, 100, 200, 500, 1000, 1500, 1800, 2000, 2200, and 3000
If these exact values are not available, you may substitute series or parallel
combinations of other values.
Entering Calibration Mode
When you are in calibration mode, the calibration step number appears in the Bipolar
display. The value(s) associated with each calibration step appears in the Cut and Coag
displays. You use the Cut and Coag up () or down () arrow buttons to adjust the
displayed values.
1. If the generator is off, turn it on.
If you removed and/or replaced the battery, alarm number 212 will appear in the Cut
display the first time you turn on the generator. Turn off, then turn on the generator
to clear the number. Then calibrate the generator.
If any other number appears, calibrate the generator before taking the appropriate
action to respond to the number.
2. To enter calibration mode, simultaneously press the Recall, Pure, and Low (Desiccate)
buttons.
The first calibration step number (C1) appears in the Bipolar display. For instructions
on completing this step, refer to the first calibration step, Verifying the Generator
Data on page 5-28.
If an error occurs during calibration, an alarm number will appear in the Cut display. Note
the number and refer to Chapter 6, Troubleshooting.
Notice
After completing any calibration step, proceed to the next step to save the values from the
completed calibration step.
Exiting Calibration Mode
You can exit calibration mode at any time. If you want to save the new values for a
particular calibration step, press the Bipolar up arrow () button at the end of that step
before exiting calibration mode. Calibration values are saved one step at a time.
For example, if you only want to adjust the clock, first enter calibration mode and press
the Bipolar up arrow () button until the generator displays calibration step 3 (C3). Next,
change the hour and/or minute values. Then press the Bipolar up arrow () button. This
saves the changes for calibration step 3 (C3) and displays the next calibration step (C4).
After saving the changes, turn off the generator to exit calibration mode.
If you do not want to save the changes you make for a particular calibration step, turn off
the generator [do not press the Bipolar up arrow () button at the end of the step]. The
values that were in effect before you started that step remain in effect.
To save the changes for the present calibration step, press the bipolar up arrow () button
to display the next calibration step number. Then, turn off the generator to exit
calibration mode.
or
To exit calibration mode without saving the changes for the present calibration step, turn
off the generator.
Verifying the Generator Data
In calibration step 1 (C1), you verify the generator model number and the software
version numbers for the main and feedback microcontrollers. You cannot adjust these
values:
• The generator code (model number) for the Force FX Electrosurgical Generator C
generator is 4.
• The software version number for the main microcontroller may vary among
manufacturing lots, but should never be 0.
• The software version number for the feedback microcontroller may vary among
manufacturing lots, but should never be 0.
If one or more of the values that appear as you complete this calibration step are not
correct, call the Covidien Service Center.
1. In the Bipolar display, verify that the calibration step number is C1.
2. In the Cut display, verify that the generator model number is 4.
3. In the Coag display, verify that the main microcontroller version number is a value
other than 0.
4. To display the feedback microcontroller version number, press the Med (Fulgurate)
button. Verify that the version number in the Coag display is a value other than 0.
5. To proceed to the next calibration step, press the Bipolar up arrow () button.
At the end of each calibration step, press the Bipolar up arrow () button to advance
to the next step. If you need to return to a previous calibration step, press the Bipolar
down arrow () button.
The Bipolar up arrow () button cycles through the calibration steps. For example,
when you reach step C10 and press the Bipolar up arrow () button, the generator
returns to step C1.
Dear Sir,
I replaced the battery but the problem 212 is still appearing. I dont know who can I set the date and time?
Please inform. Also please send me service/instruction manual. I will be grateful.
I have set the date, year and time.. and press the up button to save, but when i come back to check the date and time the setting does not stay, the 3 volt lithium battery is ok, check voltage and strength of the battery by glowing a led, bat is ok, but setting does get saved and error 212 and 216 persist, please help us sir.
Online auction of modern high-end laboratory and biopharma research equipment from the recent bankruptcy of Spexis AG in Allschwil near Basel, Switzerland. Start bidding today! closes on June 4. info@dome-auctions.com..+31(0)20 808 2880
ValleyLab Force FX Error code 212
September 04, 2013 10:03
Hello Dear Friends,
I have a problem in ValleyLab force FX electro-surgical unit. Cutting display shows Error code 212. I will be grateful if some one can inform me that how this problem can be solved.
Best regards,
Nasir.
to rate and post a comment
Riaan Bouwer
re: ValleyLab Force FX Error code 212
September 05, 2013 07:56
Code 212 is normaly internal battery. Did you replace the internal battery? If you did, go into service mode and set the time and date.
to rate and post a comment
Endomaster
re: re: ValleyLab Force FX Error code 212
September 06, 2013 08:13
Thank you so much dear Riaan. I will do the same and will inform you accordingly.
Regards,
Nasir.
Email: dmesf04@yahoo.com
to rate and post a comment
Endomaster
re: re: ValleyLab Force FX Error code 212
September 06, 2013 08:13
Please also inform me how I will go into service mode of ValleyLab force FX?
Regards,
Nasir.
to rate and post a comment
Endomaster
re: re: ValleyLab Force FX Error code 212
October 09, 2013 08:36
Dear Sir,
I replaced the battery but the problem 212 is still appearing. I dont know who can I set the date and time?
Please inform.
Regards,
Nasir.
to rate and post a comment
Abdul Awal
re: re: re: ValleyLab Force FX Error code 212
June 03, 2024 11:08
Hello,,,
Have you any solution rignt now
If you have please sheair your experience
to rate and post a comment
GHUFRAN Ullah
re: re: ValleyLab Force FX Error code 212
July 12, 2015 02:19
how i will go to service mode in Valley Lab force FX,
to rate and post a comment
Ramesh
re: re: ValleyLab Force FX Error code 212
January 27, 2023 10:23
Dear Riaan Bouwer ,
My Valleylab Force 40 Showing error 9 ...That is Checksum error in EEPROM.
plz guide how to correct it.
to rate and post a comment
William Clarke
re: ValleyLab Force FX Error code 212
September 05, 2013 07:56
What is your email address? I'll send you the manual.
to rate and post a comment
Endomaster
re: re: ValleyLab Force FX Error code 212
September 06, 2013 08:13
Dear William Clarke.
Thanks for your helping. My email is bellow.
dmesf04@yahoo.com
Regards,
Nasir.
to rate and post a comment
Endomaster
re: re: ValleyLab Force FX Error code 212
October 09, 2013 08:36
Dear Sir,
I replaced the battery but the problem 212 is still appearing. I dont know who can I set the date and time?
I need service/instruction manual.
My email is dmesf04@yahoo.com
Regards,
Nasir.
to rate and post a comment
GHUFRAN Ullah
re: re: ValleyLab Force FX Error code 212
July 12, 2015 02:19
Dear William clarke,please email me the manual because i have problem how to set date and time,
email id:ghufranjani@gmail.com
to rate and post a comment
James Allen
re: ValleyLab Force FX Error code 212
September 05, 2013 07:56
212 Generator not calibrated Refer to Calibrating the Generator on page 5-26.
Calibrating the Generator
There are eight calibration steps. During calibration you verify information specific to the
generator, adjust the date, and adjust the clock. You also adjust the REM circuit and
several values, or factors, that ensure the proper operation of the generator.
The calibration steps and the values you can adjust are summarized below. Certain values
are not adjustable but must be verified.
Notice
After completing any calibration step, proceed to the next step to save the values from the
completed calibration step.
Calibration Step and Description Adjustable?
1. Generator data
Generator model number
Master microcontroller software version
Feedback microcontroller software version
No (verify value)
No (verify value)
No (verify value)
2. Calendar
Month
Day of the month
Year
Yes
Yes
Yes
3. Clock
Hour
Minute
Yes
Yes
4. REM oscillator frequency and impedance Yes
5. Current sense gain factor Yes
6. Voltage sense gain factor Yes
7. Reactance gain factor Yes
8. ECON factor Yes
Preparing for Calibration
Equipment required
• Bipolar footswitch and monopolar footswitch
• Two small test cables (less than 24" [61 cm] long) with banana plugs
• REM plug (modified)
• Resistor substitution box
• Oscilloscope
• True RMS voltmeter (such as the Fluke 8920A or equivalent)
• Current transformer (such as the Pearson Model 411 or equivalent)
• Noninductive power resistors (such as a Dale NH-250) with these values (in): 10,
20, 30, 100, 200, 500, 1000, 1500, 1800, 2000, 2200, and 3000
If these exact values are not available, you may substitute series or parallel
combinations of other values.
Entering Calibration Mode
When you are in calibration mode, the calibration step number appears in the Bipolar
display. The value(s) associated with each calibration step appears in the Cut and Coag
displays. You use the Cut and Coag up () or down () arrow buttons to adjust the
displayed values.
1. If the generator is off, turn it on.
If you removed and/or replaced the battery, alarm number 212 will appear in the Cut
display the first time you turn on the generator. Turn off, then turn on the generator
to clear the number. Then calibrate the generator.
If any other number appears, calibrate the generator before taking the appropriate
action to respond to the number.
2. To enter calibration mode, simultaneously press the Recall, Pure, and Low (Desiccate)
buttons.
The first calibration step number (C1) appears in the Bipolar display. For instructions
on completing this step, refer to the first calibration step, Verifying the Generator
Data on page 5-28.
If an error occurs during calibration, an alarm number will appear in the Cut display. Note
the number and refer to Chapter 6, Troubleshooting.
Notice
After completing any calibration step, proceed to the next step to save the values from the
completed calibration step.
Exiting Calibration Mode
You can exit calibration mode at any time. If you want to save the new values for a
particular calibration step, press the Bipolar up arrow () button at the end of that step
before exiting calibration mode. Calibration values are saved one step at a time.
For example, if you only want to adjust the clock, first enter calibration mode and press
the Bipolar up arrow () button until the generator displays calibration step 3 (C3). Next,
change the hour and/or minute values. Then press the Bipolar up arrow () button. This
saves the changes for calibration step 3 (C3) and displays the next calibration step (C4).
After saving the changes, turn off the generator to exit calibration mode.
If you do not want to save the changes you make for a particular calibration step, turn off
the generator [do not press the Bipolar up arrow () button at the end of the step]. The
values that were in effect before you started that step remain in effect.
To save the changes for the present calibration step, press the bipolar up arrow () button
to display the next calibration step number. Then, turn off the generator to exit
calibration mode.
or
To exit calibration mode without saving the changes for the present calibration step, turn
off the generator.
Verifying the Generator Data
In calibration step 1 (C1), you verify the generator model number and the software
version numbers for the main and feedback microcontrollers. You cannot adjust these
values:
• The generator code (model number) for the Force FX Electrosurgical Generator C
generator is 4.
• The software version number for the main microcontroller may vary among
manufacturing lots, but should never be 0.
• The software version number for the feedback microcontroller may vary among
manufacturing lots, but should never be 0.
If one or more of the values that appear as you complete this calibration step are not
correct, call the Covidien Service Center.
1. In the Bipolar display, verify that the calibration step number is C1.
2. In the Cut display, verify that the generator model number is 4.
3. In the Coag display, verify that the main microcontroller version number is a value
other than 0.
4. To display the feedback microcontroller version number, press the Med (Fulgurate)
button. Verify that the version number in the Coag display is a value other than 0.
5. To proceed to the next calibration step, press the Bipolar up arrow () button.
At the end of each calibration step, press the Bipolar up arrow () button to advance
to the next step. If you need to return to a previous calibration step, press the Bipolar
down arrow () button.
The Bipolar up arrow () button cycles through the calibration steps. For example,
when you reach step C10 and press the Bipolar up arrow () button, the generator
returns to step C1.
to rate and post a comment
Endomaster
re: re: ValleyLab Force FX Error code 212
September 06, 2013 08:13
Dear James Allen,
Thank you so much for your kind information. I will try according your information.
Best regards,
Nasir.
to rate and post a comment
Endomaster
re: re: ValleyLab Force FX Error code 212
October 09, 2013 08:36
Dear Sir,
I replaced the battery but the problem 212 is still appearing. I dont know who can I set the date and time?
Please inform. Also please send me service/instruction manual. I will be grateful.
Regards,
Nasir.
to rate and post a comment
Jeetu Das
re: re: ValleyLab Force FX Error code 212
November 24, 2015 04:26
Dear James Sir,
I have set the date, year and time.. and press the up button to save, but when i come back to check the date and time the setting does not stay, the 3 volt lithium battery is ok, check voltage and strength of the battery by glowing a led, bat is ok, but setting does get saved and error 212 and 216 persist, please help us sir.
Regards,
Jeetu Das
to rate and post a comment
remzi özay
re: re: ValleyLab Force FX Error code 212
February 02, 2018 03:54
Hello there
How can I enter the calibration section of this device's calibration section?
to rate and post a comment
remzi özay
re: re: ValleyLab Force FX Error code 212
February 02, 2018 03:54
Hello there
How can I enter the calibration section of this device's calibration section?
to rate and post a comment
Abdul Awal
re: re: ValleyLab Force FX Error code 212
December 20, 2023 10:33
Thanks for your information
What's the reason 224 error
to rate and post a comment
Abdul Awal
re: re: ValleyLab Force FX Error code 212
May 30, 2024 04:02
Dear sir ,
After set date and time save press up button and power off and on again but 212 error don't gone
Have am i any mistak
Plese help me
to rate and post a comment
NAGARAJU Y R
re: ValleyLab Force FX Error code 212
July 09, 2015 01:43
How to solve 74L7 error code displaying in CUT nad COAG display of Force Fx model of Valleylab electrosurgical machine
to rate and post a comment