GHUFRAN Ullah

ValleyLab Force FX c

June 08, 2015 01:08

Dear friends,
I have a electrosurgical unit of ValleyLab Force FX c having error 160 some time also give 161 please help me.Thanks

Log inor Register

to rate and post a comment

(66)
(8)
(4)

William Clarke

re: ValleyLab Force FX c

June 09, 2015 08:46

Send me your email address and I'll send you the service manual. wlcbmt@aol.com

Log inor Register

to rate and post a comment

(61)
(24)

ariel dulay

re: re: ValleyLab Force FX c

June 15, 2015 08:32

If you can replace the components of the heat sinks on both the left and right hand sides in addition to the relay K2 that will cut short your investigation, make sure to replace all of them at the same time (no need to replace the resistor on the right hand side heat sink).

Log inor Register

to rate and post a comment

(61)
(24)

ariel dulay

re: ValleyLab Force FX c

June 15, 2015 08:32

If you can replace the components of the heat sinks on both the left and right hand sides in addition to the relay K2 that will cut short your investigation, make sure to replace all of them at the same time (no need to replace the resistor on the right hand side heat sink).

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Red Kingdom Wholesale Surplus Auctions July 24th

Red Kingdom Wholesale closeout auctions take place on DOTmed.com. Bid today on 115 lots of medical equipment. Onsite or virtual inspection available.

stats
advertisement

Événements de Vente aux Enchères à Venir

C2 Management
TIMED Ending Soon
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (229 lots)