Dear All
Spiral Toshiba CT shows the following error during the high kv scan (135 kv and ubove)
Error in XC
Code[0x62640000]
Dev [0x00010002] pno[1]
Any body explain the a.m. error.
yes I am, most high KV XC errors are caused due to a arc my first step is verify the HV cables the HV grease on these systems have a issue with drying out and breaking down if these ends are good then I would recommend running the XC maintenance program to figure out if it is the transformer, cable or tube.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Borhan Kalash
Toshiba XC error
February 14, 2014 05:36
Dear All
Spiral Toshiba CT shows the following error during the high kv scan (135 kv and ubove)
Error in XC
Code[0x62640000]
Dev [0x00010002] pno[1]
Any body explain the a.m. error.
to rate and post a comment
Lenny Place
re: Toshiba XC error
February 15, 2014 10:23
have you repacked the HV coupling gel
to rate and post a comment
Borhan Kalash
re: re: Toshiba XC error
February 16, 2014 05:19
Dear Lenny
thanks
Do you mean HV isolater grease?
Do this error means ARC?
thanks in advance
to rate and post a comment
Lenny Place
re: re: re: Toshiba XC error
February 17, 2014 10:05
yes I am, most high KV XC errors are caused due to a arc my first step is verify the HV cables the HV grease on these systems have a issue with drying out and breaking down if these ends are good then I would recommend running the XC maintenance program to figure out if it is the transformer, cable or tube.
to rate and post a comment