Toshiba XC error

February 14, 2014 05:36

Dear All
Spiral Toshiba CT shows the following error during the high kv scan (135 kv and ubove)
Error in XC
Code[0x62640000]
Dev [0x00010002] pno[1]
Any body explain the a.m. error.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Toshiba XC error

February 15, 2014 10:23

have you repacked the HV coupling gel

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: re: Toshiba XC error

February 16, 2014 05:19

Dear Lenny
thanks
Do you mean HV isolater grease?
Do this error means ARC?
thanks in advance

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Toshiba XC error

February 17, 2014 10:05

yes I am, most high KV XC errors are caused due to a arc my first step is verify the HV cables the HV grease on these systems have a issue with drying out and breaking down if these ends are good then I would recommend running the XC maintenance program to figure out if it is the transformer, cable or tube.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)