Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

system aborting scan gives error message

Tags: CT

(2)

Chidi Agunonu

system aborting scan gives error message

June 02, 2025 06:01

Am working on a GE Revolution Maxima that is aborting scan it gives the following error message:
1. Rail temperature out of tolerance
Sensor: 7
Temperature: 25.7
Limits: 37. 39.
position ID: 6-left 7-center 8-right
2. Detector Module Temperature out of tolerance
Detector Module: 1
Detector module temperature: 26.69
Limits: 37 42
what could be wrong? please help its urgent

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: system aborting scan gives error message

June 06, 2025 09:54

Looks like your rail temperature is low.
Could be an issue with the rail or modules.
You have 6 detector modules and the system is reporting module #1 as out of tolerance....its temperature is low
Contact me to troubleshoot this

Log inor Register

to rate and post a comment

(2)

Chidi Agunonu

re: re: system aborting scan gives error message

June 10, 2025 10:12

so why is the temperature going low does it mean the heater is not working?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)