Pir Muhammad ilyas

Sonos 5500 error DFD28-0001

August 24, 2015 04:28

I have a 5500 that won't boot up period. The upper screen is on but has no image period. The lower screens don't display anything.

The Back processor displays DFD28-0001

~~~~~~~~~~~~~~~~~~ From the manual ~~~~~~~~~~~~~~~~~~~~

xFD28-xxxx Indicates that the processor ROM failed core test. Four digits
of additional information indicates the ROM that failed.

Where the leading “x” can be either a K (Keyprocessor), D (Scan Converter), or
E (Scanner), representing the keyprocessor that has failed.

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

The 3.6 V battery was completely dead, so we replaced it but it didn't help the problem.

Has there been a Rom failure? Or did it lose the BIOS (or some other memory) setup?

We're not sure where to proceed from here. ???



Any help would be appreciated.



Thanks

P M Ilyas

Log inor Register

to rate and post a comment

(2681)
(674)
(21)

Lenny Place

re: Sonos 5500 error DFD28-0001

August 26, 2015 08:00

do you have the presets on disk so you can restore them

1 Press the SETUP button on the operator control panel.
2 Select SERVICE FLOPPY on the right-hand touch screen.
3 Select RESTORE PRESETS and then follow the on-screen instructions.
4 Remove the customer backup diskette from the disk drive and store it in a safe
place where it can be used if restoration again becomes necessary.

1 person liked this
Log inor Register

to rate and post a comment

(7)

Pir Muhammad ilyas

re: re: Sonos 5500 error DFD28-0001

August 29, 2015 12:06

Dear sir,
I have no Floppy disk. is there any alternate solution. how and from where i can find the required floppy disks. looking for kind reply
regards

Log inor Register

to rate and post a comment

(7)

Pir Muhammad ilyas

re: re: Sonos 5500 error DFD28-0001

August 29, 2015 12:06

Dear sir,
I have no Floppy disk. is there any alternate solution. how and from where i can find the required floppy disks. looking for kind reply
regards

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Cbrx2sales Surplus Auctions July 15th

Cbrx2sales closeout auctions take place on DOTmed.com. Bid today on 125 lots of medical equipment. Onsite or virtual inspection available.

stats
advertisement

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (501 lots)
Cbrx2sales
TIMED
Jul 15, 2026 (124 lots)
PBC Medical Solutions
TIMED
Jul 17, 2026 (11 lots)
C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (136 lots)