Somatom emotion 6
August 21, 2016 10:59
I have a problem. FATAL RAC ERROR...emotion 6 (2003)
to rate and post a comment
Lenny Place
re: Somatom emotion 6
August 22, 2016 10:14
Have you checked the fuses in the D400?
to rate and post a comment
Eric Hauk
re: Somatom emotion 6
August 22, 2016 01:02
Fatal error (Rotate Anode Controller) check for power going to anode on the tube. If no power check generator/power box. If you have power replace the tube.
to rate and post a comment