i have a somatom spirit CT scan repeated crash of the SW. we reload the SW and the system starts delayed counting after 3 days of installation and syngo not boot after 12 days of installation. The system repeated this for more than ten times. We replaced the ICS, and IRS but the problem is still.
Did you replace the whole ICS and IRS? or just changed the Tower and keeping use the Disks from original Tower?
After you changed the ICS and IRS. did you reload the software?
By thw way, where are you from?
This is strictly a software issue I seriously doubt you have any issues with the gantry. The most likely cause is a file becoming too large on your C drive and eventually causing the system not to boot. Check the CSAphonebooklog.log. Usually that is the file that builds at a rapid pace. If that file is a gig or more in size delete it and make room on your C drive.
Yes, we think in that way but we do not know what file is making the problem. The booting is progressively delay every day for example it counts to 30 in the first day, 35 in the second day 45 in the third day and so on till it reach 100 and crash. i will check the file and inform you. thanks dear.
Look through that log folder. And scan the rest of the C drive for something large. System has to have 10gig free too boot. I have a procedure to make the partition larger if need be. But I guarantee you something is too large. Customer could also be saving jpegs to that drive as well. Who knows. But poke around long enough and you find the file that's growing. If you are scared to deleted it, co twct me I will let you know if it's safe to remove, it normally is.
We did not detect any so large file but Somaris folder 1.5 giga. The C is 12 giga, 7 used and 5 free and there is no progressive increase in size of the used space. i mean it is almost the same at the time of installation and at the time of crash.
Yes that is more than a database rebuild. That is sufficient. Again it's not booting because of lack of space on the C: drive. This is a known issue with several versions of Somaris software. Send me an email ehauk@interstateimaging.com
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Bio
somaris software crash
November 01, 2015 04:44
i have a somatom spirit CT scan repeated crash of the SW. we reload the SW and the system starts delayed counting after 3 days of installation and syngo not boot after 12 days of installation. The system repeated this for more than ten times. We replaced the ICS, and IRS but the problem is still.
to rate and post a comment
Mario E Sarduy La Paz
re: somaris software crash
November 02, 2015 10:10
Hi. Which versión of software do you have? Did you try with other full set discs for do the installation?
to rate and post a comment
Bio
re: re: somaris software crash
November 02, 2015 03:18
SW VB27C
Syngo CT 2006C2
No I did not try other set discs. How can i get them?
to rate and post a comment
Mario E Sarduy La Paz
re: re: re: somaris software crash
November 03, 2015 08:23
Please send me your email o reply to me At mario.sarduy@gmail.com for more instructions or about the software
to rate and post a comment
Mario E Sarduy La Paz
re: re: re: somaris software crash
November 03, 2015 08:24
Hi. Mail me At mario.sarduy@gmail.com for more instructions and about the software
to rate and post a comment
Eric Hauk
re: somaris software crash
November 02, 2015 10:10
What version software are you using? You can reply to me at haukeye75@yahoo.com if you wish
to rate and post a comment
Mario E Sarduy La Paz
re: re: somaris software crash
November 03, 2015 08:24
Sorry I wrote you by error
to rate and post a comment
Lin Dong
re: somaris software crash
November 02, 2015 10:10
please check you UPS, and Ground.
maybe Tube Arcing?
to rate and post a comment
Bio
re: re: somaris software crash
November 02, 2015 03:18
We have replaced the UPS and checked the ground.
No tube arcing.
to rate and post a comment
Lin Dong
re: re: re: somaris software crash
November 03, 2015 08:23
Did you replace the whole ICS and IRS? or just changed the Tower and keeping use the Disks from original Tower?
After you changed the ICS and IRS. did you reload the software?
By thw way, where are you from?
to rate and post a comment
Bio
re: re: re: re: somaris software crash
November 04, 2015 08:23
We repalced the hard disks.
yes we reloaded the SW for both ICS and IRS
to rate and post a comment
Eric Hauk
re: somaris software crash
November 03, 2015 08:23
This is strictly a software issue I seriously doubt you have any issues with the gantry. The most likely cause is a file becoming too large on your C drive and eventually causing the system not to boot. Check the CSAphonebooklog.log. Usually that is the file that builds at a rapid pace. If that file is a gig or more in size delete it and make room on your C drive.
to rate and post a comment
Bio
re: re: somaris software crash
November 04, 2015 08:23
Yes, we think in that way but we do not know what file is making the problem. The booting is progressively delay every day for example it counts to 30 in the first day, 35 in the second day 45 in the third day and so on till it reach 100 and crash. i will check the file and inform you. thanks dear.
to rate and post a comment
Bio
re: re: somaris software crash
November 04, 2015 09:23
We checked the file CSAphonebooklog.log but it seems with reasonable size 4 mega.
to rate and post a comment
Eric Hauk
re: re: re: somaris software crash
November 04, 2015 01:47
Look through that log folder. And scan the rest of the C drive for something large. System has to have 10gig free too boot. I have a procedure to make the partition larger if need be. But I guarantee you something is too large. Customer could also be saving jpegs to that drive as well. Who knows. But poke around long enough and you find the file that's growing. If you are scared to deleted it, co twct me I will let you know if it's safe to remove, it normally is.
to rate and post a comment
Bio
re: re: re: re: somaris software crash
November 09, 2015 02:44
We did not detect any so large file but Somaris folder 1.5 giga. The C is 12 giga, 7 used and 5 free and there is no progressive increase in size of the used space. i mean it is almost the same at the time of installation and at the time of crash.
to rate and post a comment
Gurhan Tahtali
re: somaris software crash
November 07, 2015 01:53
Database can be the cause.Rebuild the database.Rebuid the database occasionally for maintenance.
to rate and post a comment
Bio
re: re: somaris software crash
November 09, 2015 02:44
When we reload the SW, we format the two hard disks. Is it not enough?
to rate and post a comment
Eric Hauk
re: re: re: somaris software crash
November 09, 2015 03:09
Yes that is more than a database rebuild. That is sufficient. Again it's not booting because of lack of space on the C: drive. This is a known issue with several versions of Somaris software. Send me an email ehauk@interstateimaging.com
to rate and post a comment