Siremobil compact errors

(21)
(1)

Claudio Balangione

Siremobil compact errors

February 13, 2012 05:11

Dear friends, I need some help about the following errors in a Siremobil Compact (don't have the error list):4021----4024---2006---and 5014. Thankyou in advance. Claudio Balangione

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

Re: Siremobil compact errors

February 17, 2012 09:46

Sounds like a bad memoskop.

Error E4021 Memoskop does not answer
Possible trouble: Host repeats a task 3 times, error E4021 is then entered.
Possible help: Check connection cables to the Memoskop.
Possibly Memoskop or D1 defective..


Error E4024 Memoskop initialization timeout
Possible trouble: Host registered a timeout of the Memoskop during the initialization of the system. Waiting time for the Memoskop reporting exceeded.
Possible help: Check Memoskop.
Check connection cable to the Memoskop.
Possibly Memoskop or D1 defective.

Error E2006 Memoskop, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Memoskop.
Possibly Memoskop or D1 defective.

Error F5014 Gen. Sirephos max oil pressure exceeded
Possible trouble: The oil pressure switch on the Sirephos has responded.
Possible help: Temperature of the Sirephos too high.
Check oil pressure switch.
Check cables to the oil pressure switch.
In the case of defective oil pressure switch, replace the Sirephos.
Monitoring of the oil pressure switch defective (board D1).

1 person liked this
Log inor Register

to rate and post a comment

(21)
(1)

Claudio Balangione

Re: Siremobil compact errors

February 20, 2012 10:16

Thank you again Waine. I will check those items. Claudio

Log inor Register

to rate and post a comment

(334)
(37)
(60)

Gary Benitez

re: Re: Siremobil compact errors

June 05, 2019 08:49

Hi Claudio,

Did you fix your problem?

I am having the same issues except for one.

4021, 4024, 2006.

I don't have the 5014.

Please email me at gary@crownmedicalintl.com

Thanks in advance,

Log inor Register

to rate and post a comment

(33)
(3)

Joao Carlso

re: Re: Siremobil compact errors

November 06, 2015 02:29

Dear friends, I need some help about the following error in a Siremobil Compact (don't have the error list): error 7150

Thanks

Log inor Register

to rate and post a comment

dark mackerel

re: re: Re: Siremobil compact errors

November 13, 2015 04:18

7150 means the slot diaghragm rotation index position was not found during initialization

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

General Biomedical Service - New & Refurbished Ventilators

GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449

stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)