Over 300 Virginia Auctions End Today - Bid Now Over 300 Total Lots Up For Auction at Three Locations - Over 100 California Auctions End Tomorrow 07/22 - Bid Now, KS 07/24, CA 07/31
Hi,
i have a problem for siemens somatom plus 4.
when we turn on the machine later 30 second .geting the message on screen. *can't open input device keyboard not present.useing tty for input and output*.
how can we resolve this problem.
The whole idea is the communication between host
& console is through fibre optic cable(Pls Check).The problem could be at the host board itself or at CIB connector.you may check power supply in the console box & check for any possible loose connection.These checks are to be made provided the keyboard is not faulty.
Selcuk Akin
Siemens Somatom plus 4
September 05, 2007 04:01
Hi,
i have a problem for siemens somatom plus 4.
when we turn on the machine later 30 second .geting the message on screen. *can't open input device keyboard not present.useing tty for input and output*.
how can we resolve this problem.
Best regarts
to rate and post a comment
Aydogan Polat
Re: Siemens Somatom plus 4
September 06, 2007 07:56
As I told you yesterday, if the keyboard gives the same message an a other system, just replace the keyboard.
to rate and post a comment
gustavo fiochi
Re: Siemens Somatom plus 4
September 06, 2007 04:28
the problem is the keyboard, or a loose connection of it at the computer cpu
to rate and post a comment
Sanjeev Kumar
Re: Siemens Somatom plus 4
September 08, 2007 12:00
The whole idea is the communication between host
& console is through fibre optic cable(Pls Check).The problem could be at the host board itself or at CIB connector.you may check power supply in the console box & check for any possible loose connection.These checks are to be made provided the keyboard is not faulty.
to rate and post a comment
Abel Cardona
Re: Siemens Somatom plus 4
September 08, 2007 09:10
only check the connections of fiber optic....
to rate and post a comment