Siemens Multix Pro, error code 10

February 16, 2015 12:35

X -ray table down slowly and shows Error 10. Can you help me? I thing a I need electric diagram.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Siemens Multix Pro, error code 10

February 17, 2015 10:22

Emergency stop triggered, release emergency switch. Safety circuit triggered. Check
longitudinal safety endswitches S41, S42 in 3D- TOP stand or S1, S2 in Multix Pro and
connection contactor K1.

send me an email address I will send you the manual