Miguel Chavira

serimobil 2000

October 16, 2016 01:13

Hi! i have a trouble with a c arm siemens serimobil 2000, when i turn on the system it gave me a error 32, does any body knows what is this?

Log inor Register

to rate and post a comment

DMEDS

re: serimobil 2000

October 17, 2016 10:21

Dear ,

Do you mean error 0032
it mean
0032 GEN-CURVE not allowed s no longer allowed
Cause : generator curve i
Active: during radiation Possible trouble / help : change to another curve

if you need complete manuals include the schematic diagram contact me,

Regards

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: serimobil 2000

October 17, 2016 10:21

From the manual,
S3 bus faulty or I/O bus fault
32 wait for message: Host inquires of console about configuration and waits for acknowledge message.
Display duration max. 5 sec.

Log inor Register

to rate and post a comment

(222)
(74)
(5)

Marko Shevchuk

re: serimobil 2000

October 17, 2016 10:25

Hi Miguel!

1. Error message: S3 bus faulty or I/O bus fault
2. Initialization is continued; since, however, communication (with S3 bus or I/0 bus) was impossible, further Init error messages are
shown on the display. Unit stops at Init 39.
The following unit initialization messages appear on the display of the SM2000.
3. The inquiries to messages 30 to 34 are active only in switch positions 0 and 5 of switch D6 (S3C bus) S1. In all other switch positions,
these component acknowledgements are not interrogated.

30 wait for message: Host inquires of generator about configuration and waits for acknowledge message.
Display duration max. 4 sec.
31 wait for message: Host waits for bright actual telegrams from central TV unit. Display duration max. 20 sec.
32 wait for message: Host inquires of console about configuration and waits for acknowledge message.
Display duration max. 5 sec.
33 wait for message: Host interrogates of central TV control unit about configuration and waits for
acknowledge message. Display duration max. 5 sec.
34 wait for message: Host inquires of memory about configuration and waits for acknowledge message.
Display duration max. 10 sec.

After the message 34 has been signaled, the PTW area dose meter is checked; however, only in switch position 0 of switch D6.S1. If the
calibration factor stored in the EE Prom differs from the determined value by more than 10%, the area dose meter is taken from the unit
configuration and the error message 00007 is entered in the fault storage. After run up, no m Gy/s display appears any
longer on the display. Readjustment or repair of the area dose meter is required.
All info about your error.
Regards,
Marko

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: serimobil 2000

October 17, 2016 10:36

Is it just 32 or 0032
32 wait for message: Host inquires of console about configuration
and waits for acknowledge message.
Display duration max. 5 sec

0032 GEN-CURVE not allowed
Cause : generator curve is no longer allowed
Active: during radiation
Possible trouble / help : change to another curve

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)