Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Senographe 800t problem

(30)
(1)

Francis Lienedes Ayson

Senographe 800t problem

May 20, 2008 11:22

I have a problem on a senographe 800t mammo, the console only beeps at one second interval after power is supplied on its mains input. The on/off button on its power supply board has no response with the unit.

Log inor Register

to rate and post a comment

(46)
(6)

Daniel Fernandez

Re: Senographe 800t problem

June 03, 2008 07:29

Hi Francis,
Did you check the DC/DC converter located on the right hand side of the cabinet. Power supply BD 300-PL9. It is a small box placed in the midle of the PCB. I think it converts 5 DC volts into 15 dc volts.
If it is correct,then you can check the LVPS 300 PL7, it is a comercial power supply. You can find it easily on the web.
I hope this helps you.

Log inor Register

to rate and post a comment