Alessandro Lazzari

SCT 4800 TF trouble : data AQ error

November 18, 2004 10:03

I can do the CR scan properly but during normal scan (rotation mode) I got at the end : DATA AQ ERROR ! then system go to Ready mode and no image is displayed.
Any suggestion is strongly appreciated.

Log inor Register

to rate and post a comment

Ralph Childs

Re: SCT 4800 TF trouble : data AQ error

November 18, 2004 12:22

Alessandro,

Your problem is related to rotational das pulses from the gantry. You need to check the encoder and also the input signal to the das trigger.

You can reach me at scanner1@eim.ae

Log inor Register

to rate and post a comment

(1)

Munkhtuul Togooch

re: Re: SCT 4800 TF trouble : data AQ error

March 11, 2014 08:05

Dear Ralph, Im from Mongolia, our hospital has same problem as Alessandro. How did you fix it in 2004. Sorry for asking after 10 years. Hope that you can help me!

Log inor Register

to rate and post a comment

(50)
(3)
(3)

Iftikhar Khan

Re: SCT 4800 TF trouble : data AQ error

November 19, 2004 12:09

most probably it is due to your rotational ckty.
it could be your encoder, or a rusty switch on drum. you can diagnose the defective switch it by rotating the gantry with local control box.

u can contact me on iftekhar786@hotmail.com

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: SCT 4800 TF trouble : data AQ error

November 21, 2004 11:48

hello:

if the CR image is ok, the QV boards are ok, and detector too.

i think that you must run the software for check data acquisition under rotation, to see how send info into movement.

the das cable would be broken (some lines), or encoder could be bad too.

regards,

carlos

Log inor Register

to rate and post a comment

(20)
(3)
(1)

Edwin Rosado

Re: SCT 4800 TF trouble : data AQ error

November 23, 2004 06:38

hi:
if you send me your email, i send you some instruction about how test troubleshoting this problem. your situation is in the encoder area.

Log inor Register

to rate and post a comment