Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Question

Tags: Radiology - R/F

(73)
(6)

Alaa Mohammad

Question

February 19, 2024 12:43

Hi,
If the High Tension input to the x ray tube
(Va and Vc) are not equal, and the machine(its a CArm)
is working and the image is good untill 70KV,
and it stops exposure(fluoro) for more than 70kv,
if I could overcome the protection and forced
the machine to expose for more than 70kv,
what could be the dangerous problem happen?
Thanks in advance
Alaa
exposure

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: Question

February 21, 2024 10:55

Most of generators have a feedback comparator for Kv+ against Kv-. When discrepancy are about 10%, UNBALANCED Kv signal is active.

Of course, you can override the warning, but isn't good idea, because you put one leg of inverter to deliver far more current than the other.

Even more, if you have a resonant circuit for Hv generation, you can destroy whole power stage.

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Question

February 21, 2024 10:56

Is the best way to burn the power inverter

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)