I have a problem with the table of a Philips Brilliance 64 Ct.
At the moment of passing the key to start the gantry and after turning on the table in a few seconds or minutes, the table remains locked, losing its mechanical movements, observing Registration Error # 33489129 S_HMC_COUCH_24V_IS_OFF.
After the failure, UP / DOWN movements can be performed by service.
Evaluating the failure, I observe that the power supply to the table falls, leaving the gantry on, and a part of the table is left with power.
I would like to know who can supply me with the schematic circuit of the table or the Service Manual for this Ct.