i have a problem whit a u.s. medison sa-600.
after the preventive service, the image have artefact, but not all time, somebody can help me

Log inor Register

to rate and post a comment

(55)
(1)
(1)

gustavo fiochi

Re: problem whit u.s. medison sa-600

March 30, 2005 02:54

maybe you have the display or another board with poor connection. try readjust it
what do other people think about it: if something is working well, don't touch it!!

Log inor Register

to rate and post a comment

Jim Carr

Re: problem whit u.s. medison sa-600

March 30, 2005 04:15

Sounds like it may be a problem with a probe. Is it with all transducers or only one?

Log inor Register

to rate and post a comment

(5)

ismael ceballos

Re: problem whit u.s. medison sa-600

March 31, 2005 04:01

they have only one probe but this artefact no see all time, i think the probe is good, what do you think,

sorry but my english not is good

Log inor Register

to rate and post a comment

(161)
(20)
(4)

Miguel Chavira

Re: problem whit u.s. medison sa-600

April 01, 2005 12:38

hi ! will you check in the power supply the 5V this voltage is for echoes of transducer .if this voltage raise that 5.3 o more you will regulate to 5.0 v or minor observe the image...

Log inor Register

to rate and post a comment

(5)

ismael ceballos

Re: problem whit u.s. medison sa-600

April 06, 2005 10:07

thanks to all, now is work well, was a poor contact.