Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Polydoros IT Error 45

Tags: Radiology - R/F

sohail aftab

Polydoros IT Error 45

May 05, 2012 04:22

Dears all help me. i have a problem in D500 PCB, siemens model(Polydoros IT) generator according to Error 045 i want help regarding this problem.

Err045: High Voltage Cable Configuration.

Description:
• Configuration of the high voltage cable length must be performed.
Possible Causes of Error:
• System configuration not correct.
• D500 defective.
Corrective Measures:
• Correctly configure the high voltage cable lengths
• If the error occurs again following reconfiguration, replace the D500.

i also check all thinks according to Error Description all is OK with out D500 Please help me how to conform D500 is not proper working.
I'm waiting,
supremeditech@yahoo.com Type your question or comment here...

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats
advertisement

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)