Selcuk Akin
Picker PQ CT Problem
September 17, 2007 04:29
Hi,
We have a Picker PQ CT spiral CT machine that produced a “1017 Bad Starter/No Power-DC rails less than 400VDC”.When we changed the “High Speed Starter” machine positioned “ready”, but during “warm up”(6th and 7th shooting) produced “Error 1011 and 1012”(Error Code 1011:ANA:Cathode Arc-Bad Tube/Multi/Cables. Error Code 1011: Anode Arc-Bad Tube/Multi/Cables.1012:Anode over V – Bad Multi/Cntrl.1012:Cathode over V – Bad Multi/Cntrl.1012:Tube arc…as well as “Error 1016:BAD STARTER->50amps in rails” error.
When we preheated the tube, we had several images, but then 20 amps fuses in the High Speed Starter had blown up, giving the error code 1016.
Please help us to localize this problem and the methods and steps required to solve it.
Waiting for your mail.
Best regards,
Selcuk AKIN
Note: You can send your email to this service address
service@anatoliamedical.com
to rate and post a comment
Ximena Lopez
Re: Picker PQ CT Problem
September 17, 2007 04:11
AKIN,
lOOKS LIKE YOU HAVE A FAULTY X RAY TUBE´S ROTOR, MY SUSGESTION IS TO DO A "NO LOAD TEST", TO PROBE IF ITS THE TUBE.
REGARDS
RICARDO
to rate and post a comment
Dave Adams
Re: Picker PQ CT Problem
September 17, 2007 09:24
Every time you get an error in the X-ray system, the rotor is braked and stopped. Starting and stopping the rotor so many times is likely what blew up the starter to begin with. You need to probably get a new tube, and yes do the no -load test first. But it sounds as you have a bad tube, and nobody wants to buy one. I recommend the Varian replacement, as they have a full replacement warranty, none of the pro-rated excuse for a poorly made tube like Dunlee.
Plus their tube sells thousands less money.
-Dave
to rate and post a comment
Sergio Leonardo Barral
Re: Picker PQ CT Problem
September 18, 2007 11:16
I think it is not so easy. I never got fuses blown due to a stuck rotor, and this happens a lot. No load test will give positive answers only if your Xray tube fails at minimum current. If you can make an exposure at any voltage @ 30 mA, it is very likely no load test won't help much.
to rate and post a comment
Alvaro Moreira
Re: Picker PQ CT Problem
September 20, 2007 08:14
1016 and 1017 codes are HSS related errors, even when caused by a bad tube.
No load test won´t work if you have a faulty HSS.
If you have a oil flow problem you can bypass the flow switch (in HSS) for testing purposes.
Try No Load test at 130KV and 150mA. If it runs, you have a bad xray tube.
to rate and post a comment
Kenneth Francik
Re: Picker PQ CT Problem
September 20, 2007 03:33
I had a similar problem and found a bad cable from the starter to the drives interface board.....cable l7805.Maybe it's worth checking
to rate and post a comment
Kenneth Francik
Re: Picker PQ CT Problem
September 21, 2007 07:13
The cable is L 7805 not 17805.Good luck.
Ken
to rate and post a comment
Rajoo
re: Re: Picker PQ CT Problem
March 12, 2014 01:53
Mr Kenneth francik . I am facing simillar problem . Please help me I, could nt find this cable. where this cable is located
to rate and post a comment
Rajoo
re: re: Re: Picker PQ CT Problem
April 14, 2014 08:52
I havefound that the jointer switchs which joints Tube and HSS are faulty.The switches were removed and the cables were solded directly and problem was over
to rate and post a comment
Jose Morillo
Re: Picker PQ CT Problem
October 09, 2007 08:25
Before alll....
just check income voltage. Ac voltaje less 452 volts may be the only cause. If it, use and isolation multitap transformer to adapt it. You must have 480 VAC balanced