Picker PQ 1012 Error

Jane Hall

Picker PQ 1012 Error

April 17, 2007 12:36

HV Generator KV Balance Error ? Anode & Cathode KV diff ? How can we fix it?

Log inor Register

to rate and post a comment

(93)
(7)

Edward Yan

Re: Picker PQ 1012 Error

April 17, 2007 02:25

Most likely you need a new x-ray tube replacement.

Edward Yan, B.S.E.E.

Log inor Register

to rate and post a comment

(55)
(1)
(1)

gustavo fiochi

Re: Picker PQ 1012 Error

April 17, 2007 04:09

look at the erp or erl. if you are having ANA, most probably it is the tube. how many shoots does it have?

Log inor Register

to rate and post a comment

(47)
(2)
(8)

Sergio Leonardo Barral

Re: Picker PQ 1012 Error

April 18, 2007 11:33

kV balance errors normally are born in defective multipliers, more scarcely in cabling or analog board. The problem is to determine which multiplier is the faulty one. If you have the Picker Troubleshooting manual, follow the instructions regarding the cabling. But if you reach the point to measure the feedback you certainly can be unsure which one is the faulty multiplier because the HV generator shuts down before you can get a significative measurement even using a scope. At that point I suggest you to try the multipliers in another scanner or have spare ones and alternativelly replace them until you fix the problem.

Log inor Register

to rate and post a comment

(55)
(9)
(3)

Jim Foros

Re: Picker PQ 1012 Error

May 03, 2007 03:32

I have had this several times, it was always a multiplier.

Log inor Register

to rate and post a comment