Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

picker explorer II error E95

(174)
(8)
(31)

Hipolito Jorge Iglesias

picker explorer II error E95

March 01, 2012 09:58

DEar Friends
if someone has had the problem with error code E95 with a portable x ray machine Picker explorere II and want to share with me the troubleshooting to fix this faillure any help will be very appreciatted.
annorys1971@att.net.mx

Log inor Register

to rate and post a comment

(102)
(13)

Joel Grullon

Re: picker explorer II error E95

March 07, 2012 06:04

that error mean the kv is low, usually that coming from battery bad, conect at gnd and (b7+) meter recorder max/min ie fluke and set the meter for min. then make exposure if the voltaje get down at least 15 percent change the bat

best regards
joel

Log inor Register

to rate and post a comment

(174)
(8)
(31)

Hipolito Jorge Iglesias

Re: picker explorer II error E95

March 07, 2012 10:32

Dear Joel

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)