Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Picker Eclipse 6U 9U interface problem

Tags: MRI, PICKER

Posted by an Anonymous User

Picker Eclipse 6U 9U interface problem

June 01, 2012 08:53

Hi to everybody;
I have a problem on Picker Eclipse MRI.I am writing Xterm fault line ;

-scan pmc6u9u:3 pciFindDevice failed;could not find 6U to 9U interface PMC PCI bus.
-scan bpbInitLib:3 Unsupported access interrupt detected ;expected timeout/no response type.
-scan S_feiInitLib_ILLEGAL_BACKPLANE_BUS_ACCESS

system logging to scan,recon,pap0,pap1 and system loading PGR but not logging to FEI server.
Do you have any experience about this topic?
Best Regards

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

re: Picker Eclipse 6U 9U interface problem

June 02, 2012 03:38

Hi,
PGR, scan, recon, pap are parts of 6U section. FEI - Front End Interface is a board from 9U. From first view it is like a not working 9U section (Power Supply?). Check the LEDs on the front panel and leds on the 9U boards. Possible faulty PS.

Log inor Register

to rate and post a comment

(10)

(1)

Sanchez Navarro

re: re: Picker Eclipse 6U 9U interface problem

June 04, 2012 09:25

Thanks for reply dear Selwesiuk,9U PS leds are on,also i replaced a new one 18V PS but problem is still there.I replaced 9U backplane board,Data Engine,TR interface but problem is didnt change..

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

re: re: re: Picker Eclipse 6U 9U interface problem

June 04, 2012 02:46

PS provides +5v, -18v and +18vdc.
9U clock is derived from Exciter board (under the 9u unit). Check the PS for Exciter/Receiver section - leds should be on - on the left side of the boards.
Make photos of the any visible leds and 6u and 9u boards and send me to analyze (wselwesiuk@synektik.com.pl)

Log inor Register

to rate and post a comment

(10)

(1)

Sanchez Navarro

re: re: re: re: Picker Eclipse 6U 9U interface problem

June 05, 2012 09:26

All scan/recon PS leds are on Mr Selwesiuk,but i didint check to exciter cables connection.If you want to see photos of Scan/recon,i will send.
Thanks

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

re: re: re: re: re: Picker Eclipse 6U 9U interface problem

June 06, 2012 09:10

OK, send me