PICKER 5000 CT(CRX-400),when warm up and scan image ,1006 come out
ERROR CODE 1006,TRACER:8
XSC+ STATUS:Xray software.ALE won't go away.Xray tube may not be discharge KV below 30KV or bad multiplier (HV Module if 400R) or bad hvg control bd.
How to do next?Thank you!
Hi
It can be produced for several causes. First, be sure that slip rings are clean and ok. you can run test "ring" from cdddiag sub directory. Choose speed 5 and let it go. If any error is present it will be showed.
If slip rings are ok, discart that XSC software is ok. Reload software using "gxd xsc" command. Don't touch anything during loading software.
If problem persist, use "iqout" program and type "xe" to see status of XSC, follow instructions showed.
Probably you will need to use "No LOAD" test to identify failing module.
A question. DO you obtain any image or the error appear just beggining the exposure?
Feel free to write for further instruction if you want or need
Thank you for your reply!
When you proceed pliot scan 130KV 30MA,it will display 1006 error on Touch Panel after exposed,"READY" turn off, and also you could get pilot image.if you want to plan ,you must turn the key again to "READY" light and scan image.I find a regular pattern ,That is related to KV,
low KV,less error!I could change the scan parameter to 120KV,But the pilot parameter can not be changed.I want to solve the problem completely.
Best regards,
fumingyou
Hi
If you are sure the slip ring are clean and ok, you must run a NO LOAD test to find whisch HV component of the HVG is failing. It can be Tube, inverters, output box or multipliers. NO LOAD test will help you to identify the cause.
Do you tried modifying mA? If problem appear more with more mA it can be inverters. If just change with KV it can be multipliers or tube. Verify oil level inside the HV cables connectors (multipliers side, do not try to unscrew it from tube)
Hi,I proceed "NO LOAD TEST",120KV,130KV,140KV,NO ERROR.I doubt the reason be the tube.Next month,I will find a tube to confirm.Thank you for your help.
Hi Abdul
I am confused perhaps for the english, it's clear it is not the native languaje for both os us.
i need to be sure I understood:
1.- You have error 1006 when try to do warmup and you can not do it.
2.- if you try to take pilot, the system move and sound like it do the pilot but not ray (sure? may be there is Xray but there is not image)
3 also have error 2005
2005 - Programmable Interrupt (PPI) error:
The following occur through programmable interrupts:
1. Programmed Position
2. Data Acquisition Complete Couch is NOT moving!
3. Data Grabber
4. Scan Position Valid
If one of these could not be enabled because of some other process already waiting for that interrupt, or did not occur within an expected time this error message is generated. Check the XSC STAT: 5th nibble, Bit 1 (Exposure armed). Sync signal uses slip ring 7 & 8. Also, DREQ grounded - Data cable cut!
i am sure its the large filament on tube your tube is bad you need to replace tube i had this problem also you need to check igpt on invertor it can be lost from tube arching
thanks
Hi Sayat. Please do you can say me the weather conditions to produce 1006? I had worked on CT's at high altitude and extremely cold (3 to 10°C) without 1006 appear. The humidity is extremely high there but without condensation. I plan to return in a couple of months and want to know if there is something I can do to prevent it to appear
Thanks in advance
i have been servicing Picker Cts for more than 14 years.
this is very rare trouble.
and i faced it just one time and solved it.
pls contact me at bkalash71@yahoo.com
mingyou fu
PICKER 5000 CT ERROR CODE 1006
December 21, 2012 11:56
PICKER 5000 CT(CRX-400),when warm up and scan image ,1006 come out
ERROR CODE 1006,TRACER:8
XSC+ STATUS:Xray software.ALE won't go away.Xray tube may not be discharge KV below 30KV or bad multiplier (HV Module if 400R) or bad hvg control bd.
How to do next?Thank you!
to rate and post a comment
Jose Morillo
re: PICKER 5000 CT ERROR CODE 1006
December 23, 2012 12:41
Hi
It can be produced for several causes. First, be sure that slip rings are clean and ok. you can run test "ring" from cdddiag sub directory. Choose speed 5 and let it go. If any error is present it will be showed.
If slip rings are ok, discart that XSC software is ok. Reload software using "gxd xsc" command. Don't touch anything during loading software.
If problem persist, use "iqout" program and type "xe" to see status of XSC, follow instructions showed.
Probably you will need to use "No LOAD" test to identify failing module.
A question. DO you obtain any image or the error appear just beggining the exposure?
Feel free to write for further instruction if you want or need
Best regards
jose
to rate and post a comment
mingyou fu
re: re: PICKER 5000 CT ERROR CODE 1006
December 24, 2012 10:11
Thank you for your reply!
When you proceed pliot scan 130KV 30MA,it will display 1006 error on Touch Panel after exposed,"READY" turn off, and also you could get pilot image.if you want to plan ,you must turn the key again to "READY" light and scan image.I find a regular pattern ,That is related to KV,
low KV,less error!I could change the scan parameter to 120KV,But the pilot parameter can not be changed.I want to solve the problem completely.
Best regards,
fumingyou
to rate and post a comment
Jose Morillo
re: re: re: PICKER 5000 CT ERROR CODE 1006
December 25, 2012 09:36
Hi
If you are sure the slip ring are clean and ok, you must run a NO LOAD test to find whisch HV component of the HVG is failing. It can be Tube, inverters, output box or multipliers. NO LOAD test will help you to identify the cause.
Do you tried modifying mA? If problem appear more with more mA it can be inverters. If just change with KV it can be multipliers or tube. Verify oil level inside the HV cables connectors (multipliers side, do not try to unscrew it from tube)
to rate and post a comment
mingyou fu
re: re: re: re: PICKER 5000 CT ERROR CODE 1006
December 28, 2012 09:40
Hi,I proceed "NO LOAD TEST",120KV,130KV,140KV,NO ERROR.I doubt the reason be the tube.Next month,I will find a tube to confirm.Thank you for your help.
to rate and post a comment
Jose Morillo
re: re: re: re: re: PICKER 5000 CT ERROR CODE 1006
December 28, 2012 04:06
Sometimes a bas tube cobdition is miss interpretated by software. Recently I changed all HV chain to find tube was BAD with error 1009....
to rate and post a comment
Mazher saeed
re: re: re: re: re: re: PICKER 5000 CT ERROR CODE 1006
May 23, 2013 05:52
Pikcer IQ premier Extra
XRAY ERROR,System in interlock
when i want to make warm up
i checked all switches please guide me that how can i solve this problme
thnx
Also the num 2 is error2005 that
when system warm up |
then i want to take pilot couch move silently back but no exposure or radiation takes place
to rate and post a comment
Jose Morillo
re: re: re: re: re: re: re: PICKER 5000 CT ERROR CODE 1006
May 24, 2013 10:11
Hi Abdul
I am confused perhaps for the english, it's clear it is not the native languaje for both os us.
i need to be sure I understood:
1.- You have error 1006 when try to do warmup and you can not do it.
2.- if you try to take pilot, the system move and sound like it do the pilot but not ray (sure? may be there is Xray but there is not image)
3 also have error 2005
2005 - Programmable Interrupt (PPI) error:
The following occur through programmable interrupts:
1. Programmed Position
2. Data Acquisition Complete Couch is NOT moving!
3. Data Grabber
4. Scan Position Valid
If one of these could not be enabled because of some other process already waiting for that interrupt, or did not occur within an expected time this error message is generated. Check the XSC STAT: 5th nibble, Bit 1 (Exposure armed). Sync signal uses slip ring 7 & 8. Also, DREQ grounded - Data cable cut!
to rate and post a comment
Mazher saeed
re: re: re: re: re: re: re: re: PICKER 5000 CT ERROR CODE 1006
July 02, 2013 05:03
Thank you
SYSTEM IN INTERLOCK was solved
that data cable on Rotor controle card was not installed in correct position
i unpluged it and installed in right position system in interlock
go away
also evry fault cleard
to rate and post a comment
Jose Morillo
re: re: re: re: re: re: re: PICKER 5000 CT ERROR CODE 1006
May 24, 2013 10:11
write direct to e at jgmorillo@gmail.com
to rate and post a comment
Jim Shami
re: re: PICKER 5000 CT ERROR CODE 1006
December 25, 2012 09:34
i am sure its the large filament on tube your tube is bad you need to replace tube i had this problem also you need to check igpt on invertor it can be lost from tube arching
thanks
to rate and post a comment
Jose Morillo
re: re: re: PICKER 5000 CT ERROR CODE 1006
December 27, 2012 03:16
You can easily be sure about it using iqout program and selecting between filaments.
to rate and post a comment
mingyou fu
re: re: re: PICKER 5000 CT ERROR CODE 1006
December 28, 2012 09:41
I use "manexp" to change the filamaent,without any change, report 1006 error,too.
to rate and post a comment
Sayat Karapetyan
re: PICKER 5000 CT ERROR CODE 1006
December 23, 2012 12:59
Hi,
More than ten reasons that give error 1006. But mostly it's coming from a dirty slip rings and the low temperature of the room Gantry.
to rate and post a comment
Jose Morillo
re: re: PICKER 5000 CT ERROR CODE 1006
December 25, 2012 09:36
Hi Sayat. Please do you can say me the weather conditions to produce 1006? I had worked on CT's at high altitude and extremely cold (3 to 10°C) without 1006 appear. The humidity is extremely high there but without condensation. I plan to return in a couple of months and want to know if there is something I can do to prevent it to appear
Thanks in advance
to rate and post a comment
mingyou fu
re: re: PICKER 5000 CT ERROR CODE 1006
December 28, 2012 09:51
Temperature is kept between 20--24℃ .That is OK all the time.
to rate and post a comment
Jose Morillo
re: re: re: PICKER 5000 CT ERROR CODE 1006
December 28, 2012 04:06
This level of temperature is a little hot for the equipment. I use to set it to 18°C or less
to rate and post a comment
Borhan Kalash
re: PICKER 5000 CT ERROR CODE 1006
December 31, 2012 09:47
dear mingyou,
i have been servicing Picker Cts for more than 14 years.
this is very rare trouble.
and i faced it just one time and solved it.
pls contact me at bkalash71@yahoo.com
Regards
Borhan
to rate and post a comment
Domingo Perez
re: PICKER 5000 CT ERROR CODE 1006
May 24, 2013 10:15
XRAY_SOFTWARE_ERROR
Possible BAD HOST SEQUENCE, RELAY BD, DIB, RCB, STARTER, SLIP RING
to rate and post a comment
mingyou fu
re: PICKER 5000 CT ERROR CODE 1006
May 24, 2013 08:47
All friends,We changed the tube.That is working well.I think there is some trouble in the tube insert .Thank you for your help!
to rate and post a comment
Borhan Kalash
re: re: PICKER 5000 CT ERROR CODE 1006
May 25, 2013 09:10
Dear mingyou,
If u want I can help u fix the tube
it happend with us and we fix it
email me at bkalash71@yahoo.com
to rate and post a comment