Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Philips sonos error code

Tags: Ultrasound, PHILIPS

(18)
(3)

Bharat Ramchandani

Philips sonos error code

June 22, 2023 12:04

Hello. Can anyone help urgently me with a philips sonos 7500 error code EF79E. Machine was working fine then on next start up came up with this code and cannot work around it. Many thanks for your help.

Log inor Register

to rate and post a comment

(25)
(3)

heretohelp

re: Philips sonos error code

June 23, 2023 10:41

Here is the error decoded. E=Scanner portion, F=Conditional error, 79E=Most probable cause is the Distribution Board or the Transducer.

Log inor Register

to rate and post a comment

(18)
(3)

Bharat Ramchandani

re: re: Philips sonos error code

August 07, 2023 11:34

Hi, Many thanks for your reply. I swapped out the distribution board but couldnt find a 3D distribution board for the 7500, so thought Id try a 5500 distribution board, hoping it would be compatible. It allowed me to boot up normally but gave a message saying that the 3 D probe would not be funcional. However, on attempting to plug in non-3D probes like the S8 or S4, it would bring up anoter messafge saying to reconnect the probe and press reset. This would lead to a freeze situation. So I wanted to ask is the 5500 distribution boarn not compatible with the 7500? Alos i was getting an 0E04 error on the LED indicator on the processor graphics board which, accoring to the 5500 service manual, indicates a T-Cine-SW RAM failure . Any idea on how to proceed. Can't seem to find a replacement 3D distribution board 77110-20440 anywhere online, but there are the 5500 distrribution boards available, which is why i thought I would try it and didn't mind loosing the 3D probe functionality.

Log inor Register

to rate and post a comment

(95)
(18)
(2)

Paulo Ferreira

re: Philips sonos error code

June 23, 2023 10:42

Hello,
Check distribution board or transducer. Remove transducer and try again.

Log inor Register

to rate and post a comment

(18)
(3)

Bharat Ramchandani

re: Philips sonos error code

August 07, 2023 11:29

Hi thanks for the replies. Reconnecting transducer does not work. Removed distribution board but couldn't source a 7500 distribution board so tried a 55oo distribution board as I thought it may be compatible. It worked to allow normal boot up and made it to operating screen with no transducers connected. It did give a warning that no 3D probes could be used, but still allowed me to get to main screen. However on attempting to connect any transducers it would give an error message to "reconnect transducer and press reset" after which the system would attempt to reboot but then freeze up with the monitor not even turing on.
Anyone know if i can make the 5500 distribution board work in the 7500. Was the only option I had since they are 5500 DB availabe to purchase second hand but no 7000 DBs.
Will appreciate any advice given.
Also scanned 7500 DB for any burned components but nothing looks abnormal. Anyone has any experience with repairing these boards and any ideas on what could be the issue?
Many thanks again.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Medical Equipment Superstore Surplus Auctions July 14th

Medical Equipment Superstore closeout auctions take place on DOTmed.com. Bid today on 276 lots of medical equipment. Items can be inspected in person or via video call. Contact Rick at +1 (215) 669-3913 or info@medicalequipmentsuperstore.com

stats
advertisement

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)