If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare
Philips DuoDignost erro 3231003
January 01, 2016 10:26
Gentelmen ,
Error 3231003 is displayed .
When freecasate technique is selected , colimator size not controlled ; x ray image is 5cm width .
Please details & steps to solve the problem
Regards
to rate and post a comment
Fastone Oscar RUGOTWA
re: Philips DuoDignost erro 3231003
January 02, 2016 02:35
Hi my friend
This 3231003 is a fatal error
Description:
GEO:SEC:Mech.or electr.error:RM Motor moved too.
Possiple cause:
Belt broken,check the drive
Regards Fastone Oscar
CEO FANESTECH (T) LTD.
to rate and post a comment
Fastone Oscar RUGOTWA
re: Philips DuoDignost erro 3231003
January 02, 2016 02:36
Hi,
Where are you from, am engineer Fastone from Tanzania,are you a technical personel or user?
If you need an Eng for rectifying the problem i can help my friend.
However if you are a technician i can guide what to do for the machine.
Regards Eng.Fastone Oscar.
to rate and post a comment
necdet sahin
re: Philips DuoDignost erro 3231003
January 02, 2016 02:37
hello
Write the e-mail.
necdet şahin
necexray@hotmail.com
to rate and post a comment
Galal
re: re: Philips DuoDignost erro 3231003
January 03, 2016 10:19
thanks alot
to rate and post a comment
umberto sciunnach
re: Philips DuoDignost erro 3231003
January 02, 2016 02:39
Hello
On service manual there is written the type of error.
If I remember is security error and you can found simply.
On the base of remote control there is a plastic box.
Read error on the manual and I can help you how eliminate.
My mail umbertosc1957@gmail.com
My mobile for whatsapp +393891529020.
Regards
Umberto
to rate and post a comment
Galal
re: re: Philips DuoDignost erro 3231003
January 03, 2016 10:19
thanks a lot
to rate and post a comment
Fastone Oscar RUGOTWA
re: re: re: Philips DuoDignost erro 3231003
January 04, 2016 10:09
Hi
As i know about this error,you can connect pc kit on the console and through you pc if x scope if istalled,you can open and go to error log,and check the possible cause.
If not open the serial changer on the table and move out the whole serial changer,then check the Belt for Right motor and fix it,the machine ill work.
My mail :fastone.oscar@yahoo.com
Regards Fastone
to rate and post a comment
Mostafa Lotfy
re: Philips DuoDignost erro 3231003
May 20, 2016 04:36
Check RM Encoder disc and adjust with LM Encoder disc , to be with in range as visual.
But the high possibility "flat cable" inside SCE.
to rate and post a comment