OR 35 MAGNET QUESTION
July 02, 2007 11:05
FOR QUENCH OR 35 MAGNET WHEN İNSTALLİNG DSL PROBE DO WE READ A SHORT CIRCUIT? NOTE: ERROR CODE 105 IS READ N THE DİSPLAY AT THİS TİME
to rate and post a comment
Iosif Sogolov
Re: OR 35 MAGNET QUESTION
July 03, 2007 07:29
is the switch heater on?
Marc Fessler
July 03, 2007 09:29
You probably know by now, but you should read less then 1 ohm. Only do this on magnet without field.