Hi friends, this unit goes to stand alone after booting and when exposure switches are activated it displays error107.
I have checked interface cables between master board D702&D740panel board. What can I do?
I also need the software for this equipt for calibration.
Please help, my email; tubomie@gmail.com

Log inor Register

to rate and post a comment

(40)
(6)
(1)

Santosh Kumar Tiwari

re: Nova 3000-Error 107

April 10, 2014 08:57

Dear Tim
If No Data transmission takes place between the Deck & Master between the first 20 second after power on the power deck switches to the so called "Stand Alone" mode.
The Key Board Can be operated normally. If the exposure release button is operated in this mode the board maintained error message is displayed on the display.
The Cause of this error can be an interruption on the serial interface to the master. For this reason this connection must be checked exactly.
Thanks
Santosh K Tiwari