Dear seniors,
As my Positioner NVRAM battery expired, i changed it with new one the i flash it with my CPU firmware Figaro 7.1
my target is MVME 162-150A
My email k-okos@hotmail.es
Thank you
At first you should Configure your MVME board and load the new NVRAM. (Once the Battery fails the board will be set to default. You will have to set it according to the GE Service manual)
Secondly you should Calibrate the Positioner and load the settings to the NVRAM.
MIGUEL ANGEL ROCHA
MVME162-510A NVRAM E Positioner help needed
November 01, 2016 10:56
Dear seniors,
As my Positioner NVRAM battery expired, i changed it with new one the i flash it with my CPU firmware Figaro 7.1
my target is MVME 162-150A
My email k-okos@hotmail.es
Thank you
to rate and post a comment
R GOPAL
re: MVME162-510A NVRAM E Positioner help needed
November 03, 2016 10:38
Hi,
At first you should Configure your MVME board and load the new NVRAM. (Once the Battery fails the board will be set to default. You will have to set it according to the GE Service manual)
Secondly you should Calibrate the Positioner and load the settings to the NVRAM.
R.Gopal
aparnaelectromech@gmail.com
to rate and post a comment
Sergiy Shurayev
re: re: MVME162-510A NVRAM E Positioner help needed
May 11, 2017 09:17
Hi,
could you tell me, please Direction # of GE manual you talked about.
BR,
Sergiy Shurayev.
to rate and post a comment