Over 150 Missouri Auctions End Today - Bid Now

Millennium MPR Detector Calibration

Tags: Nuclear / Molecular Imaging

(20)
(8)

Danilo Abdo

Millennium MPR Detector Calibration

December 20, 2016 02:20

Hello All!

I get here a instalation from ˜cold˜, it is I mean, from zero.. Cristal with no tubes, nothing.
So, I install all tubes ( 48 tubes ) and Matrix, IN board, All cables, all necessary, whith a GX200. ( software 4.5 )

So, what I can`t find a solution, is because many tubes ( some times 21 tubes ) becomes out of range, very high. All this tubes already are in low on adjustment.

Can someone give me a help on this?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Millennium MPR Detector Calibration

December 21, 2016 10:48

Have not worked that much on the GE line but on Siemens and Adac I would be verifying my HV power supply and lv supplies. Run the proactive PMT diagnostics and check results

1 person liked this
Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: re: Millennium MPR Detector Calibration

December 28, 2016 09:01

Thanks Lenny, but the power supply is ok.. I change it 2 times to make sure.. and I can run proactive PMT is not have any pre map on Genie Systems....

Log inor Register

to rate and post a comment

(9)
(1)

BENITO MONTOYA VARELA

re: Millennium MPR Detector Calibration

December 27, 2016 10:15

Danilo , on the case of PM´t some times is written the H.V. PMT check this and si if it is the correct, if you have the same ,please disconnect the 1/2 of the PMTubes and check the others, ,and do the same with the other ,a badPMT can cause your problem , . Also check the type of PMT using in this systems.Suerte

Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: re: Millennium MPR Detector Calibration

December 28, 2016 09:01

Hello Benito

What you mean with disconnect 1/2 tubes? PMTUNE will works on half tubes on detector? The PH from tubes is about minimum 160-200mv , and one or 2 tubes on 230~ mv.

Can you help me on it?

thanks!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)