if you have made the calibration procedure and it has not remedied the issue try cleaning the temperature probe with a mild abrasive pad.
Also check or replace the bellows. if all this fails replace the temp probe.
it would help alot if you can tell me the conditions at the time of failure i.e. chamber pressure. and temp. and condition of door gaskets (door and dam)
Regards,
Hi Richard i really apreciated your help. I put a new below, i clean the temp probe, i made the calibration procedure, the problem is just when the pressure is around 28 psi ( 198 kpa) and temp is around 120 then show ERROR CODE 5, door gaskets are okey. See like the temp probe, what do you think?
Ok, I am assuming the 120 Degree is in Celsius.and we know that there are no pressure leaks, start a sterilization cycle, With an multimeter monitor the resistance on the temp probe as it heats up make sure that the it does not go open at any time during the cycle.
Also i am starting to suspect a faulty thermostat at the bottom of the chamber test it in the same fashion .
give me a call at 305 978-8031 perhaps i can be of more help over the phone...if I don't pick up the phone leave me a call back number.
Hi Richard, for sure the temp is in Celsius, for sure not pressure leaks, i will check the sensor doesn´t open. Very good point of the termostat i will double check, i really appreciated your time and help, my number is +52 5550681413, thanks a lot
MArcio, thanks a lot for your answer, well my friend
i put a new bellow, i already order a new temp sensor and a new termostat, just when i will get it i put and then i tell you if works
thanks
francisco
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
Francisco Jaramillo
MIDMARK M 11 ERROR CODE 005
May 28, 2013 06:40
some one has experience with this. I have service manual, i made calibration process doesnt work
to rate and post a comment
Richard Fernandez
re: MIDMARK M 11 ERROR CODE 005
May 29, 2013 09:41
if you have made the calibration procedure and it has not remedied the issue try cleaning the temperature probe with a mild abrasive pad.
Also check or replace the bellows. if all this fails replace the temp probe.
it would help alot if you can tell me the conditions at the time of failure i.e. chamber pressure. and temp. and condition of door gaskets (door and dam)
Regards,
Rich
to rate and post a comment
Francisco Jaramillo
re: re: MIDMARK M 11 ERROR CODE 005
May 30, 2013 09:47
Hi Richard i really apreciated your help. I put a new below, i clean the temp probe, i made the calibration procedure, the problem is just when the pressure is around 28 psi ( 198 kpa) and temp is around 120 then show ERROR CODE 5, door gaskets are okey. See like the temp probe, what do you think?
to rate and post a comment
Richard Fernandez
re: re: re: MIDMARK M 11 ERROR CODE 005
May 31, 2013 11:13
Ok, I am assuming the 120 Degree is in Celsius.and we know that there are no pressure leaks, start a sterilization cycle, With an multimeter monitor the resistance on the temp probe as it heats up make sure that the it does not go open at any time during the cycle.
Also i am starting to suspect a faulty thermostat at the bottom of the chamber test it in the same fashion .
give me a call at 305 978-8031 perhaps i can be of more help over the phone...if I don't pick up the phone leave me a call back number.
to rate and post a comment
Francisco Jaramillo
re: re: re: re: MIDMARK M 11 ERROR CODE 005
May 31, 2013 03:22
Hi Richard, for sure the temp is in Celsius, for sure not pressure leaks, i will check the sensor doesn´t open. Very good point of the termostat i will double check, i really appreciated your time and help, my number is +52 5550681413, thanks a lot
to rate and post a comment
Marcio Braga
re: re: re: re: re: MIDMARK M 11 ERROR CODE 005
May 31, 2013 06:15
hi, friends.If you calibrated and is giving this error, probably the valve BELLOWS is dirty or clogged.steam is coming out of the trap 90C?
to rate and post a comment
Francisco Jaramillo
re: re: re: re: re: re: MIDMARK M 11 ERROR CODE 005
June 03, 2013 01:52
MArcio, thanks a lot for your answer, well my friend
i put a new bellow, i already order a new temp sensor and a new termostat, just when i will get it i put and then i tell you if works
thanks
francisco
to rate and post a comment