Over 150 Missouri Auctions End Today - Bid Now

Lunar Expert Bone Densitometer

(1)

Lewis Davies

Lunar Expert Bone Densitometer

April 06, 2005 08:25

Does anyone have any service info on this model please. The machine fails it's daily Qa check at the point where lights are tested and the XRay Tube is warming up, message displayed on monitor says failure during XRay ramp-up. The fault will go away for the rest of the day if the vertical cables encased in platic trunking at the bottom of the C-arm are wiggled and QA tests re-done. Any help appreciated
Contact lewisd@adhb.govt.nz

Log inor Register

to rate and post a comment

(10)

John Cude

Re: Lunar Expert Bone Densitometer

April 07, 2005 08:54

I take care of 2 of these units. If one of the lamps are out it will fail that portion of the test. I do have the service literature for this unit, and some parts. These units are a total mechanical nightmare. John Cude 203 380 5129

Log inor Register

to rate and post a comment

(5)

ismael ceballos

Re: Lunar Expert Bone Densitometer

April 07, 2005 01:54

maybe the lamps are bad or check the cables, check too, an arcin
luck

Log inor Register

to rate and post a comment

stats