Manuel Brugat

Lorad MIV Platinum AEC Problem

April 06, 2017 10:14

Dear Colleagues:

I am having a problem with a Lorad MIV Platinum.

In auto-time the result is higher mAs. than the one I get regularly for the Phantom that has to be submitted for ACR.
I suspected that I needed to do the GAIN adjustment for the AEC.
I attempted to adjust but even at maximum gain the voltage I get at TP7 is approximately 7.5 volts measured with the oscilloscope as per procedure. (The voltage should be approximately 9.0-9.5 volts)
Is there anyone who can help me with this problem?

Sincerely:

Manuel Brugat/Image Instrumentation, Inc.

Log inor Register

to rate and post a comment

(28)
(1)

alexander grau

re: Lorad MIV Platinum AEC Problem

April 07, 2017 10:20

shoulbe you used caibration AEC with plexiglass, my email is alexgrau1966@gmail.com, speaking spanish

Log inor Register

to rate and post a comment

(145)
(16)

Juan Antonio Ramirez Ramirez

re: Lorad MIV Platinum AEC Problem

April 07, 2017 10:21

Most times I just adjust gain master d
And offset for each kvp station. Just pressing the four switches of right side board.of course it depends on wich chisis I am using. I MEAN kodak fuji ...