Lorad M-IV system 35 error message

Tim Austin

Lorad M-IV system 35 error message

July 08, 2008 04:55

I have a lorad m-iv that is getting a system 35 error message with very small breasts. I went through the filament calibration and thought that took care of the problem, but a week later the problem resurfaced. Anyone have any ideas?

Thanks

Log inor Register

to rate and post a comment

Tim Austin

Re: Lorad M-IV system 35 error message

July 14, 2008 02:14

I also checked and adjusted the primary filament current because it was a little low. The problem again went away for a week and now has come back again, but has gotten progressively worse to the point that the error shows up anytime you try to take an exposure. Any ideas on the problem and/or possible solution would be greatly appreciated.

Log inor Register

to rate and post a comment

(2)

noel fleck

Re: Lorad M-IV system 35 error message

July 15, 2008 11:43

You probally already know this but in case you don't. Error 35= tube current is below 10% OF EXPECTED VALUE. Check tube current level,calibration,filaments, and protective diodes across filaments.

Log inor Register

to rate and post a comment