Lorad Hologic Selenia DRAPI error

Tags: Mammo, LORAD

(192)
(45)
(1)

Dante Nuevo

Lorad Hologic Selenia DRAPI error

April 23, 2014 09:13

Hello good friends,

I am having trouble with a Lorad Hologic Selenia mammo unit. I made a calibration on request of the system, once it was made I made a reboot and the system crashed. Now it shows an error saying:

Error initializing the DRAPI
There was a problem initializing the drapi.
Check for
1) a missing configuration file for the detector
2) a corrupt gain calibration or map file.

Can anyone please give an advice on how to repair this error? how can I check the config, calib and map files???

Best regards.

Dante Nuevo.

Log inor Register

to rate and post a comment

(2681)
(674)
(21)

Lenny Place

re: Lorad Hologic Selenia DRAPI error

April 24, 2014 01:12

do you have a calibration file you can restore, if so restore then perform the calibrations again

Log inor Register

to rate and post a comment

(192)
(45)
(1)

Dante Nuevo

re: re: Lorad Hologic Selenia DRAPI error

April 24, 2014 02:59

Hello Lenny, thanks for your answer.

Which are the files that I need to restore?

In case I don't have any of these files which would be the procedure?

Best Regards.

Dante Nuevo

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: Lorad Hologic Selenia DRAPI error

April 28, 2014 08:53

Hi,

You need to check the log files, check the drapi.log file.

Log inor Register

to rate and post a comment

(2)

soloana Kautar

re: re: Lorad Hologic Selenia DRAPI error

July 15, 2019 10:19

I have no files like Dr. Cfg because I made new software installation and now unable to load detector files one thing more sun pc 150p is not connected with gantry .
Kindly let me know solution.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats
advertisement

Événements de Vente aux Enchères à Venir

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (218 lots)