eduardo medina

LORAD Elite error

October 08, 2008 10:54

When turn on machine send a error # 28 Realease x rays button, but are not pressed. Even disconnect buttoms and same.I tried disconect and connect, even press a reset buttom and not work. Any idea?
Eduardo

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: LORAD Elite error

October 10, 2008 08:13

Ed:

Check in the board where the handswitch is connected.
In the service manual describe the xray butom is depresed when in mode standby.

Manuel

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

Re: LORAD Elite error

October 10, 2008 10:26

Hello Manuel, Ny customer are seeking a service manual,ASAP have it, I hope look through it.
Thank you

Log inor Register