lorad affinity error 12 nvram failed

January 28, 2015 08:39

I have a problem with a Mammo - Lorad Affinity Mammo
Model No. 3-000-5066

The Lorad Affinity is presenting an error code: Error 12 ( nvram failed during power on )
I will appreciatte any ideas, Thanks.

Log inor Register

to rate and post a comment

(54)
(7)
(9)

Deborah Breland

re: lorad affinity error 12 nvram failed

January 28, 2015 05:38

It appears that you will need to replace the NVRAM. If you email me at debi.breland@digitecmedical.com I will give you the information on the one we have in stock.

Log inor Register

to rate and post a comment

(10)
(3)

Natthakit Prongchai

re: lorad affinity error 12 nvram failed

January 29, 2015 10:03

You need to replace the NVRAM P#M4T28-BR12SH and then you need to restore backup data into the unit. If you do not have the backup data. You need to run all calibration.

Log inor Register

to rate and post a comment

(38)
(12)

surasit kruata

re: re: lorad affinity error 12 nvram failed

January 30, 2015 09:13

Thank so much k.Natthakit for your reply my email.

Log inor Register

to rate and post a comment

(67)
(3)
(1)

Lobsang Ortiz

re: re: lorad affinity error 12 nvram failed

February 01, 2015 10:24

Hi, I had the same error, I replaced the NVRAM, but I don't have the backup data, can you help me with instructions or service manual, an error 44 appears some times after x ray. Regards

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: lorad affinity error 12 nvram failed

February 01, 2015 10:26

do you measure voltage in backup cell of nvram? I have this trouble and fix it by a 3V IonLi cell

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Medical Equipment Superstore Surplus Auctions July 14th

Medical Equipment Superstore closeout auctions take place on DOTmed.com. Bid today on 276 lots of medical equipment. Items can be inspected in person or via video call. Contact Rick at +1 (215) 669-3913 or info@medicalequipmentsuperstore.com

stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)