i have problem with logiq400pro. The problem is when change the probe from 1st to 2nd (probe not problem)m/c gives error(sys error:(SHVCV-PC)error and m/c hung up. kindly give the solution,thanking you,
Hellow sir,
Thanking for imt. rosponse.But the problem is m/c hung up and no pattern when change the probe form 1st 2nd.But there is no problem when changing the probe from 1st to 3rd or 3rd to 1st. so kindly give the solution.Thanking you sir.
Rajan Alagu
logiq 400 pro problem
July 24, 2008 09:27
i have problem with logiq400pro. The problem is when change the probe from 1st to 2nd (probe not problem)m/c gives error(sys error:(SHVCV-PC)error and m/c hung up. kindly give the solution,thanking you,
to rate and post a comment
Lakshmi Narayana
Re: logiq 400 pro problem
July 24, 2008 01:31
HV voltage power supply could be problem
to rate and post a comment
umut keles
Re: logiq 400 pro problem
July 25, 2008 09:56
hi
probe select
4 X 1n4148 or power supply unit +12V and -12V
to rate and post a comment
Rajan Alagu
Re: logiq 400 pro problem
July 26, 2008 03:03
Hellow sir,
Thanking for imt. rosponse.But the problem is m/c hung up and no pattern when change the probe form 1st 2nd.But there is no problem when changing the probe from 1st to 3rd or 3rd to 1st. so kindly give the solution.Thanking you sir.
to rate and post a comment
Manuel Casanova
Re: logiq 400 pro problem
July 27, 2008 08:43
Dear sir:
Please check in the Advance service manual, and make to transducer test, I will be think that your problem is fixed to TD.
Is posible you check the transducer in other equiment.
Check the power supply 380VDC A10 HV REG ASSY.
Manuel
to rate and post a comment
Bharat sharma
Re: logiq 400 pro problem
July 29, 2008 11:02
problem in the multipluxer[probe connector ] pcb which holds the all probes ,nothing should be wrong with power supply.
to rate and post a comment
Rajan Alagu
Re: logiq 400 pro problem
August 14, 2008 09:54
Thank you sir, i will check it.
to rate and post a comment