Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

LOGIQ 400 ERROR

Tags: Ultrasound

Tomas Camargo

LOGIQ 400 ERROR

June 25, 2012 03:25

Hi everyone, I hope someone can help!

I have a logiq 400 system that is giving me the following error.
Warning(AV ABNORMAL. PCONT [M12VL1LG])any help will be appreciated.

Thanks in advance.

Log inor Register

to rate and post a comment

Gilberto Rubio Gil

re: LOGIQ 400 ERROR

June 25, 2012 05:34

probably you have a problem with the HV supply.
where are you from? I am in Guadalajara, Mexico. Maybe i can help you.

Log inor Register

to rate and post a comment

Tomas Camargo

re: re: LOGIQ 400 ERROR

June 25, 2012 09:50

Thank you Gilberto, do you think you can send me your phone number so I can call you, iif you can please send to tcamargo3@gmail.com.
Thanks in advance.

Log inor Register

to rate and post a comment

(8)
(3)
(32)

sitwat kashif

re: LOGIQ 400 ERROR

June 25, 2012 07:26

if you need service manual of L-400 contact with me ON FOLLOWING EMAIL: sitwatmedical@gmail.com

Log inor Register

to rate and post a comment

Tomas Camargo

re: re: LOGIQ 400 ERROR

June 25, 2012 09:50

Thanks men but I already have the munual.

Log inor Register

to rate and post a comment

Josip Vujnovic

re: re: re: LOGIQ 400 ERROR

June 26, 2012 09:33

Solution: UL-SOLUTION-002510

Title: UL_L400_HIGH VOLTAGE ABNORMAL PCONT_HVPWOVER



Solution Description:

H V Abnormal PCONT HVPWOVER



Resolution:

Possibly the BMRC brd
BMRC makes REFCV Control Signal for the TRDR
the REFCV controls the HV voltage.

Log inor Register

to rate and post a comment

(45)
(6)

Miguel Flores

re: LOGIQ 400 ERROR

June 26, 2012 09:41

POWER SUPPLY ASSY/PCONT ASSY/Power Control Assembly

Log inor Register

to rate and post a comment

Marcelo Antonio Correa

re: LOGIQ 400 ERROR

June 26, 2012 09:41

Hello, Tomas! Replace R101, R102, R103 on the 3 port probe connector board for 47 Ohms, 1/4 Watt resistors. Check the old ones first, they are probably tending to infinity. If is a DCONN (2-port probe board), I'm not sure, but must be R31 and R32. Hope this is helpful, God Bless. MAC.

1 person liked this
Log inor Register

to rate and post a comment

(54)
(5)

kco

re: re: LOGIQ 400 ERROR

November 22, 2016 10:08

good day Marcelo,
I have the same problem and I followed the indications that you mention, the resistors are good, the voltages of the source are within the range, I need help to solve these problems, but you can give me a tip, to solve.

1 person liked this
Log inor Register

to rate and post a comment

(28)
(7)

Madhavesh Kulkarni

re: LOGIQ 400 ERROR

June 26, 2012 09:36

Dear Tomas
pls check +-5/12/15vdc outputs on power supply assy. one/more output must b missing. you can repair it too.

Regards
Madhavesh

Log inor Register

to rate and post a comment

(97)
(4)

Vimal Prakash Tripathi

re: LOGIQ 400 ERROR

January 12, 2013 11:25

dear tomas,
did u got resolve the problem regardings.

Log inor Register

to rate and post a comment