LINES ARTIFACTS WHILE SCANNING WITH EVERY PROBE

(3)
(1)

nicola femiano

LINES ARTIFACTS WHILE SCANNING WITH EVERY PROBE

February 07, 2024 05:51

Hi everyone, I have a problem with my Ge Vivid I which I bought recently used. It worked perfectly for a month, then started showing vertical lines and artifacts in any mode with both probes. The next day everything went back to normal for another 4-5 days. When the system works well the sistem does not report any errors, while when the lines appear it reports errors in both RFI and DC offset. do you have any idea what the problem could be? Thank you

Log inor Register

to rate and post a comment

(4)
(1)
(5)

Matthew Sanchez

re: LINES ARTIFACTS WHILE SCANNING WITH EVERY PROBE

February 09, 2024 10:19

this is usually do to outside noise. The problem with noise is it is intermittent