The I.I. has no output but all voltages are there, hear ticking (arcing) from bottle, lost vacuum? Any help will be much appreciated jvanderwerf@aol.com
disconnect only the white anode cable and try again, if it still arcs, then depending on the age and manufacturer it may be possible to repair.
hope this helps.
Hello Felix, I much appreciate your response, if it still arc's with the anode disconnected what might the problem be. Note this is a reinstall and it was working prior to the move. It's a newer 33cm Siemens I.I.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
John Vanderwerf
image intensifier problem
October 24, 2009 08:23
The I.I. has no output but all voltages are there, hear ticking (arcing) from bottle, lost vacuum? Any help will be much appreciated jvanderwerf@aol.com
to rate and post a comment
Felix Donoghue
Re: image intensifier problem
October 24, 2009 11:58
disconnect only the white anode cable and try again, if it still arcs, then depending on the age and manufacturer it may be possible to repair.
hope this helps.
to rate and post a comment
John Vanderwerf
Re: image intensifier problem
October 26, 2009 09:33
Hello Felix, I much appreciate your response, if it still arc's with the anode disconnected what might the problem be. Note this is a reinstall and it was working prior to the move. It's a newer 33cm Siemens I.I.
Thanks again for your help, John
to rate and post a comment
Felix Donoghue
Re: image intensifier problem
October 28, 2009 10:47
If it still arcs then Grid 3 is gone or tube has leaked maybe damage while shifting. better replace. sorry not much help.
to rate and post a comment