Ross Corbett
IE33 error codes
June 14, 2013 01:51
Does anyone have a list of the IE33 error codes?
And if anyone knows what error code 034 is that would be helpful.
Thanks
to rate and post a comment
Rick Reed
re: IE33 error codes
June 14, 2013 05:16
Code 034
Condition
Calibration Error On The Omni Motor Controller
Description
The system has detected a failure in calibration for the motor controller.
Corrective Action
Straighten the TEE probe and disconnect and reconnect transducer.
Scanning cannot begin until calibration is successful. If the problem
persists: replace the transducer or front-end controller.