I am having a problem with an HP ImagePoint HX ultrasound. When performing transcranial studies sometimes there is a failure when using the Doppler. I get the message "An error has ocurred" EF512 - 00009060. I will appreciate any help you can provide. Thank you.
Regards,
Elias Valdes

Log inor Register

to rate and post a comment

Oral Unal

Re: HP ImagePoint

October 04, 2005 10:41

power supply out of regulation. Adjust gain pot. Check transmit power supply

Necati Unal

Log inor Register

to rate and post a comment

Jim Carr

Re: HP ImagePoint

October 04, 2005 11:55

From our systems engineer:
EF512
Scanner Fatal - Power supply regulation, Check Transmit supply. (GAIN control can cause "this error in certain setups") Does not say if TCD is involved...
Possible the xdcr could be bad...