Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

HP 43100A ERROR 2

(1)

ahmed youssef

HP 43100A ERROR 2

May 04, 2009 06:26

GOOD MORNINIG
I AM DR AHMED YOUSSEF FROM EGYPT PLEASE I NEED YOUR HELP
> I HAVE 2 HP 43100A-43110A DEFIBRELATOR
> THAT GIVE ERROR 2
> IT GIVE NO CHARGE I DO NOT KNOW WHY
> PLEASE IF YOU HAVE ANY OPARATION MANUAL
> OR YOU CAN ASK YOUT TECHNIMAN
> PLEASE
> NO ONE HERE IN EGYPT KNOW THE CAUSE
> I DO EVRY THING TO SOLVE THIS PROBLEM BUT NO WAY
> I ASK YOUR HELP....PLEASE THIS IS THE LAST HOPE
SOME ONE SAY TO ME THIS IS DEAD BATTARY BUT I HAVE OTHER UNIT THAT STILL WORK GOOD I TAKE ITS BATTARYBUT STILL THE ERRORWHAT I CAN DOPLEASE I NEED YOUR TELEPHONE NOTO SPEAK TO YOU

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

Re: HP 43100A ERROR 2

May 05, 2009 09:17

e-mail me back I can help

Log inor Register

to rate and post a comment

(39)
(4)

Hernandez Hugo

Re: HP 43100A ERROR 2

May 05, 2009 01:22

Hi, follow instrucctions please: ERR 2 is Defib charging too slow.
Place equipmernt in the service mode (press and hold SYNC button. Turn the ENERGY SELECT knob to the MONITOR position. Release SYNC button after about one second. A test wawefor should appear on the CRT).
Unplug the unit from AC power. Read de battery voltage using up-down arrows. If the battery is above 12.3 volts prooceed, otherwise allow it to charge fully for 8 hours or more.
Set energy to 360 J and press the CHARGE button. The CHARGE DONE tone should indicate charge completion in less than 10 seconds.
Maybe your problem is battery damage.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)