Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Hologic Mammography System (Loard Affinity) image problem

Tags: Mammo, HOLOGIC

(11)
(5)

yikwang

Hologic Mammography System (Loard Affinity) image problem

August 04, 2014 02:53

Dear all,
During imaging, I found the output image intermittent appear white pieces in the middle/corner edge of image.
What would be the possible root cause for this problem?

For mirror assembly, the position of mirror will affect the x-ray field area? (I try to move the mirror position and the collimate area will be different so i wonder the position will affect the x-ray field or not.)

Log inor Register

to rate and post a comment

(28)
(1)

Larry Underriner

re: Hologic Mammography System (Loard Affinity) image problem

August 06, 2014 12:10

It's been awhile, but as I recall the Mirror is supposed to pull out of the field during the exposure. My guess is that it is not pulling all the way out? You can get into the Service Menu and change the home and park positions with DAC Values. Unfortunately, I do not have a Service Manual anymore to clarify.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)