Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Hologic Exporer Problem

Tags: Bone Densitometer, HOLOGIC

(48)
(4)

Emrah Kambak

Hologic Exporer Problem

January 28, 2018 11:25

Dear Friends
I had to remove the carm of the hologic explorer qdr unit to move to another room in the hosp. After reinstallation I get error message while QC procedure right after test image completed (39 second) "test failed - set SD exceeded. Phantom out of position- absolute max or min limit exceeded- an error has occurred in a csfstream" please advice. Thank you!

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: Hologic Exporer Problem

February 05, 2018 09:36

check the x-ray alignment, from the error it seems that the x-ray reaching the detector is low.

2 people liked this
Log inor Register

to rate and post a comment

(48)
(4)

Emrah Kambak

re:re: Hologic Exporer Problem

February 05, 2018 09:51

Thank you.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Brand new OEM Philips & Siemens Parts at unbeatable prices

Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)