Hi friends
I'm working bone densitometer delphy
The system have a lot trouble, some time the calibration pass fine but some time not pass and show error that phantom out position, this happen because de arm not move for scan phantom
Y calibrate AY movement and this action no problem, now show me other error XRAY FIRING ORDER IS OUT SEQUENCE. STATE MACHINE ERROR IN TASK 0 INDEX 278
the calibration AY appear on graph mode not values like service manual, appear other character's not values numbers or letter
I think so that software is damaged
Some help me with this
Thanks
Hi,don't mess with the software, you have to fix the filter drum first. Do replace the timing belts and check with the pin if timing will be OK after their replacement.Also, remove the 3 allen screws that hold the base of this mechanism, put it aside and check inside the precolimator cup for oil leak (rempve the 2 outer screws of the round lead colimator). If you see oil, make sure that you will retighten all 8 screws of the cup using the correct torque qnd sequence.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Germni
Hologic Delphy trouble
October 16, 2023 02:19
Hi friends
I'm working bone densitometer delphy
The system have a lot trouble, some time the calibration pass fine but some time not pass and show error that phantom out position, this happen because de arm not move for scan phantom
Y calibrate AY movement and this action no problem, now show me other error XRAY FIRING ORDER IS OUT SEQUENCE. STATE MACHINE ERROR IN TASK 0 INDEX 278
the calibration AY appear on graph mode not values like service manual, appear other character's not values numbers or letter
I think so that software is damaged
Some help me with this
Thanks
to rate and post a comment
AKIS PAPAS
re: Hologic Delphy trouble
October 17, 2023 10:42
Hi,don't mess with the software, you have to fix the filter drum first. Do replace the timing belts and check with the pin if timing will be OK after their replacement.Also, remove the 3 allen screws that hold the base of this mechanism, put it aside and check inside the precolimator cup for oil leak (rempve the 2 outer screws of the round lead colimator). If you see oil, make sure that you will retighten all 8 screws of the cup using the correct torque qnd sequence.
to rate and post a comment
Germni
re: re: Hologic Delphy trouble
October 19, 2023 10:19
Thank you, I detect fail battery bios then replace battery y configuration serial and paralelo port
Now solve all trouble
to rate and post a comment
MRAdmin
re: Hologic Delphy trouble
October 18, 2023 10:18
is this a new installation of a used system? or it is an old installation.
you have many problems, and the system need calibrations and checks
to rate and post a comment
Germni
re: re: Hologic Delphy trouble
October 19, 2023 10:19
Thank you