Over 500 Total Lots Up For Auction at One Location - NJ 06/15

HITACHI APERTO

Tags: MRI

(6)

richardgomez

HITACHI APERTO

August 29, 2017 09:03

Type yourHi everyone,I have a problem with a MRI HITACHI APERTO,
I show the following error:
Scan error the resonant frequency is not found frequency lock check the MR
unit and the coil error no 12135000 prescan software.

My inspections:
1- the error exist with all the coils.
2- the magnet temp. is ok 32 degree
3- at the table coil connector the coax pins are ok

What do you think the cause and how to solve the problem?

Regards. question or comment here...

Log inor Register

to rate and post a comment

(1)
(1)
(1)

Miguel Moura

re:HITACHI APERTO

August 30, 2017 10:02

Hello Malcom.
Where are you based? Which company?
Maybe I can help with Aperto problem

1 person liked this
Log inor Register

to rate and post a comment

(6)

richardgomez

re: re:HITACHI APERTO

August 31, 2017 10:17

thanks for your answer.
The système is in dakar ;Sénégal
but the problem has be resorber.it was the frequence search

Log inor Register

to rate and post a comment

(6)

richardgomez

re: re:HITACHI APERTO

August 31, 2017 10:17

thanks for your answer.
The système is in dakar ;Sénégal
but the problem has be resorber.it was the frequence search

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re:HITACHI APERTO

August 31, 2017 10:18

Ive got almost the same problem on Airis Mate system
The problem appears usually after the scanogram at the first attempt of the 2nd sequence. After this no scano is possible . Restarting the system sometimes 1 examination can be done. No more, the same error appears.

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: HITACHI APERTO

September 02, 2017 03:07

check the receiving path, it looks that the receiving path have problem.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI Solutions to Meet Your Team’s Unique Needs

From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)