When I try to make a exposure on a GE Hispeed NXi CT an error comes up
Host:Hispeed Proc: tgp_out Error: 186518
File:: tgp_errors.c Line:312
DAS Error
Code: 30-1010-01
Error detected by: CIF
Zero DAS data was detected
sometimes it shows another error :Das data over range.
the trouble is intermittent
any recommendations
Thank you
Hello, zero data and overrage, means that the signal coming from detectors reaches down and up limits respectively. This could be a noisy signal from amplification of onevchannel, a miscontact in the signal path, or plain noise, most probably coming from power supplies.
Some action for this error: 30-1010-01
Potential Cause:
1. CAM; 2. CIF; 2. DDP
I would like to know if you can see an image when the error is present.
Normally this problem is because you have a bad channel or many bad channels.
There are a procedure when you get this error.
You can analysis the image and get a graphic for the bad channel.
The localize the channel and the bad CAM.
Send me your eMail to send you the troubleshooting for this error.
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
Borhan Kalash
Hispeed Nxi zero das
November 11, 2013 04:12
When I try to make a exposure on a GE Hispeed NXi CT an error comes up
Host:Hispeed Proc: tgp_out Error: 186518
File:: tgp_errors.c Line:312
DAS Error
Code: 30-1010-01
Error detected by: CIF
Zero DAS data was detected
sometimes it shows another error :Das data over range.
the trouble is intermittent
any recommendations
Thank you
to rate and post a comment
Bruce Hull
re: Hispeed Nxi zero das
November 11, 2013 06:31
This can be the result of a noisy or failed DAS power supply. Check all DAS power supplies for DC accuracy and AC ripple.
to rate and post a comment
Kim Schofield
re: re: Hispeed Nxi zero das
November 12, 2013 08:46
Also check the brushes
to rate and post a comment
Borhan Kalash
re: re: Hispeed Nxi zero das
January 07, 2014 04:13
thanks Brent
what is the acceptable max. ripple value?
to rate and post a comment
Carlos Duran
re: Hispeed Nxi zero das
November 12, 2013 08:45
Hello, zero data and overrage, means that the signal coming from detectors reaches down and up limits respectively. This could be a noisy signal from amplification of onevchannel, a miscontact in the signal path, or plain noise, most probably coming from power supplies.
to rate and post a comment
Borhan Kalash
re: re: Hispeed Nxi zero das
January 07, 2014 04:13
which PS is the most frequent show such trouble 5 or 15?
to rate and post a comment
Raul Elizalde
re: Hispeed Nxi zero das
November 13, 2013 06:44
Some action for this error: 30-1010-01
Potential Cause:
1. CAM; 2. CIF; 2. DDP
I would like to know if you can see an image when the error is present.
Normally this problem is because you have a bad channel or many bad channels.
There are a procedure when you get this error.
You can analysis the image and get a graphic for the bad channel.
The localize the channel and the bad CAM.
Send me your eMail to send you the troubleshooting for this error.
Good look.
to rate and post a comment
Lin Dong
re: Hispeed Nxi zero das
November 17, 2013 01:31
check the DAS transfer. das system, transmitter or receiver?
to rate and post a comment
Borhan Kalash
re: Hispeed Nxi zero das
November 23, 2013 10:32
thank u all
thank u rosa
I can not get image.
my email
bkalash71@yahoo.com
to rate and post a comment