Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Hispeed Nxi zero das

Tags: CT

(183)
(13)
(2)

Borhan Kalash

Hispeed Nxi zero das

November 11, 2013 04:12

When I try to make a exposure on a GE Hispeed NXi CT an error comes up
Host:Hispeed Proc: tgp_out Error: 186518
File:: tgp_errors.c Line:312
DAS Error
Code: 30-1010-01
Error detected by: CIF
Zero DAS data was detected

sometimes it shows another error :Das data over range.
the trouble is intermittent
any recommendations
Thank you

1 person liked this
Log inor Register

to rate and post a comment

(3)

(3)

Bruce Hull

re: Hispeed Nxi zero das

November 11, 2013 06:31

This can be the result of a noisy or failed DAS power supply. Check all DAS power supplies for DC accuracy and AC ripple.

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: re: Hispeed Nxi zero das

November 12, 2013 08:46

Also check the brushes

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: re: Hispeed Nxi zero das

January 07, 2014 04:13

thanks Brent
what is the acceptable max. ripple value?

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: Hispeed Nxi zero das

November 12, 2013 08:45

Hello, zero data and overrage, means that the signal coming from detectors reaches down and up limits respectively. This could be a noisy signal from amplification of onevchannel, a miscontact in the signal path, or plain noise, most probably coming from power supplies.

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: re: Hispeed Nxi zero das

January 07, 2014 04:13

which PS is the most frequent show such trouble 5 or 15?

Log inor Register

to rate and post a comment

(39)
(8)

Raul Elizalde

re: Hispeed Nxi zero das

November 13, 2013 06:44

Some action for this error: 30-1010-01
Potential Cause:
1. CAM; 2. CIF; 2. DDP
I would like to know if you can see an image when the error is present.
Normally this problem is because you have a bad channel or many bad channels.
There are a procedure when you get this error.
You can analysis the image and get a graphic for the bad channel.
The localize the channel and the bad CAM.
Send me your eMail to send you the troubleshooting for this error.

Good look.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: Hispeed Nxi zero das

November 17, 2013 01:31

check the DAS transfer. das system, transmitter or receiver?

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: Hispeed Nxi zero das

November 23, 2013 10:32

thank u all
thank u rosa

I can not get image.
my email
bkalash71@yahoo.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Brand new OEM Philips & Siemens Parts at unbeatable prices

Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)